The signal will always be weaker with CP light. But it will no longer be orientation dependent. See Pu Xu, Eleanor Kable, Colin J. R. Sheppard, and Guy Cox, 2010 A quasi-crystal model of collagen microstructure based on SHG microscopy. Chinese Optics Letters 8, 213-216
Do be aware that your quarter wave plate must be quarter wave for your excitation wavelength - a visible light one will not work if you are using SH / 2P excitation. (Lambda/4 at 900 nm = lambda/2 at 450nm). The analyser is less critical.
Guy
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Alberto Diaspro
Sent: Thursday, 9 August 2012 11:01 PM
To:
[hidden email]
Subject: Re: Imaging collagen with circularly polarized light
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
maybe this set-up could help
Diaspro, A., Bertolotto, M., Vergani, L., & Nicolini, C. (1991). Polarized light scattering of nucleosomes and polynucleosomes-in situ and in vitro studies. IEEE TRANSACTIONS ON BIOMEDICAL ENGINEERING, 38(7), 670-678.
Il giorno 09/ago/2012, alle ore 14:15, Michael Abanto <
[hidden email]> ha scritto:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Hi,
> Does anybody have a good protocol for imaging collagen in embedded tissue, stained with picrosirius red? I had no problem imaging collagen with linear polarization, but I wanted to try circular and found it difficult.
> I have an Olympus BX61 and a Nikon A1, as well as demo lambda plates for both. Do I need an analyzer specific for circular polarization? And do I need a linear polarizer before the circular - and how would you set that up on the Olympus?
> Any tips or help on the set-up would be appreciated,
> Mike