Dear All,
Thank you for all the help.
Final conclusion: We need the information about the laser focused volume on the stage for both two photon and one photon excitation laser, so we are going to measure the FWHM of objective PSF with the diffraction limited beads.
Best regards,


TONG Yan
Department of Biological Sciences
National University of Singapore


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox
Sent: Friday, August 31, 2012 6:48 AM
To: [hidden email]
Subject: Re: focused laser beam size on stage

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

How is that possible?

                               Guy

Optical Imaging Techniques in Cell Biology by Guy Cox  2nd edition 2012  CRC Press
     http://www.guycox.com/optical.htm
______________________________________________
Guy Cox, MA, DPhil(Oxon), Honorary Associate, Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net
 


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Pascal Weber
Sent: Friday, 31 August 2012 4:49 AM
To: [hidden email]
Subject: Re: focused laser beam size on stage

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Ok Craig thank you for these precisions. I knew that Olympus uses a driven telescope  for modulating the beam diameter before entering the scan head.
That does not change the filling of the rear back lens.