Re: Oil vs water objectives

Posted by Scot C Kuo on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Oil-vs-water-objectives-tp7579114p7579120.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I've measured it quantitatively and the performance difference flips surprisingly close to the coverslip (see supplemental info, Fisher & Kuo 2009 PNAS 106, 133-138).  For an Olympus 60x U-PlanApoS lenses, comparing 1.2 and 1.4 NA, the flip happens ~8 microns into an aqueous sample.  For fluorescence closer than ~8um, oil is brighter, whereas for objects further, water immersion is brighter.  If lenses aren't matched, then the cross-over can happen elsewhere, but the relative shapes of the curves are the same.  Oil lenses (1.4NA) will have half the brightness by ~50um.

For the information you've provided (higher NA on water lens), I'd expect the cross-over to be closer to the coverslip surface.

-- Scot

============================================================================
...............Scot C. Kuo (410) 955-4536; email:[hidden email]...............
...Director, Microscope Facility, JHU-SOM, www.hopkinsmedicine.org/micfac...
..Assoc Professor, Biomedical Engineering & Cell Biology, www.jhu.edu/cmml..


----- Original Message -----
From: Gabriel Lapointe <[hidden email]>
Date: Thursday, October 4, 2012 9:16 am
Subject: [CONFOCALMICROSCOPY] Oil vs water objectives
To: [hidden email]


> *****
>  To join, leave or search the confocal microscopy listserv, go to:
>  
>  *****
>  
>  Hi,
>  
>  I have a user who insist that using a 1,27NA water immersion
> objective is
>  brighter and would give better images than using a 1,4NA oil
> immersion. I
>  understand that deeper into the media that would be true. But, in that
>  particular case, we are talking about imaging GFP at less than 100 micron
>  away with a spinning disk.
>  
>  So, I was wondering at which distance from the coverslips do we start
>  seeing benefits of using a water immersion objective over an oil objective
>  in aqueous media.
>  
>  Thanks for your help.
>  
>  Sincerely
>  *Gabriel Lapointe, M.Sc.*
>  Lab Manager / Microscopy Specialist
>  Concordia University, Biology Department
>  7141 Sherbrooke St. West SP 534
>  MontrĂ©al QC H4B 1R6 Canada
>  [hidden email]
>  cmac.concordia.ca
>