Re: Question about deconvolution

Posted by Lutz Schaefer on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Question-about-deconvolution-tp7579203p7579210.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I do agree with most of the posts here and like to add a few more things.

From a theoretical point of view and preferably in noiseless continuous
space, there is a clear limitation of restoration of spatial frequencies up
to the optical resolution for an unconstrained linear regularized inverse
system. It will improve the FWHM of a point object, but not beyond the Airy
disk. Using a nonlinear, for example positivity constrained iterative
method, this limitation becomes lifted as frequencies beyond that limit are
allowed to enter (to understand how, there is a paper from Merserau et al.).
Especially in the widefield case, the gain in axial resolution becomes
easily apparent but only with this class of methods and with proper
background correction. From the theory, I would see no limitation in
regaining spatial frequencies beyond optical resolution. Awakening usually
strikes in the real world, where we deal with photon limited discrete
measurements as George points out. In the end it is a battle with the signal
to noise and the ill-posed-ness of the underlying problem to be solved. This
battle is lost to these factors and regularization has to be chosen to mend
noise amplification which diminishes the goal as it suppresses high
frequencies. There are older papers from Jose Conchello and Rainer Heitzmann
(sorry I don't have the references handy) who try to achieve the utmost
improvement, but with only mild success. Also Walter Carrington's paper,
mentioned from Guy lies in this category, but he was rather focusing on a
very solid rigorous mathematical concept, that would achieve this goal.

To make a long story short. Deconvolution can greatly improve the contrast
of fine structures. Some methods can in theory break the diffraction limit,
but with real data this can only be achieved at best in axial dimension of a
widefield modality.

Regards
Lutz

__________________________________
L u t z   S c h a e f e r
Sen. Scientist
Mathematical modeling / Image processing
Advanced Imaging Methodology Consultation
16-715 Doon Village Rd.
Kitchener, ON, N2P 2A2, Canada
Phone/Fax: +1 519 894 8870
Email:     [hidden email]
___________________________________


--------------------------------------------------
From: "George McNamara" <[hidden email]>
Sent: Tuesday, October 23, 2012 21:48
To: <[hidden email]>
Subject: Re: Question about deconvolution

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Confocal and multiphoton excitation fluorescence deconvolution will
> greatly benefit from customers requiring vendors to include fast,
> quantitative (precise, accurate) deconvolution on all confocal/MPEF
> systems bought/sold. GPU enabled deconvolution. If three academics can
> make GPU SIM (pretty much) realtime (and a heck of a lot faster processing
> than OMX), the hardware vendors ought to be able to do it (or license it):
>
>    Lefman J, Scott K, Stranick S. Live, video-rate super-resolution
>    microscopy using structured illumination and rapid GPU-based
>    parallel processing. Microsc Microanal. 2011 Apr;17(2):191-6. Epub
>    2011 Mar 9. PMID: 21385522
>
> If the vendors don't get it done, I hope everyone goes for do it yourself
> nanoscopes, such as:
>
>    *York* AG, Parekh SH, Dalle Nogare D, Fischer RS, Temprine K, Mione
>    M, Chitnis AB, Combs CA, *Shroff* H. Resolution doubling in live,
>    multicellular organisms via multifocal structured illumination
>    microscopy. Nat Methods. 2012 May 13;9(7):749-54. doi:
>    10.1038/nmeth.2025. PMID: 22581372
>    http://code.google.com/p/msim/
>
> Good luck with the vendors.
>
> ***
>
> As for deconvolution "breaking" the optical limit: of course not. The
> deconvolution is done on your computer, the optical part ended when the
> photons slammed into the detector(s). I go with Paul Goodwin's (Applied
> Precision) rule of thumb (or maybe pinky finger?): deconvolution can
> improve the IMAGE resolution by ~20%.
>
> There are other ways to get improvement in resolution - I am very pleased
> with the increase in IMAGE resolution of PiMP:*
>
> *
>
>    *Munck S*, Miskiewicz K, Sannerud R, Menchon SA, Jose L, Heintzmann
>    R, Verstreken P, Annaert W. Sub-diffraction imaging on standard
>    microscopes through photobleaching microscopy with non-linear
>    processing J Cell Sci. 2012 May 1;125(Pt 9):2257-66. PMID: 22357945.
>
> On the Leica SP5 inverted and Zeiss LSM710 confocal microscopes i manage
> here at the U, I use 30 nm pixel size (vs 60 nm for "standard" high
> resolution confocal images), 25 to 50 images (when I'm acquiring for a
> user, my core makes more money if image acquisition time takes longer,
> single channel (makes dealing with Fiji ImageJ easier - would be nice if
> someone took the time to develop ImageJ into a real program with user, not
> anti-user, interface), no averaging, unidirectional scanning. My SP5 has
> both standard and resonant scanners - both work (resonant is faster). Goal
> is modest photobleaching (ideally ~5% according to the paper). Since 30 nm
> pixel size is smaller than the PiMP plugin default, I use filter size =
> 1.65. My thanks to Sebastian Munck and Glen Macdonald for the
> recommendation and for putting in 16-bit output mode, which I routinely
> use. Does PiMP work? Yes. Don't take my word for it: go look at Figure 2
> of the paper and/or get the plugin from the corresponding author.
>
> George
>
>
> On 10/23/2012 9:23 PM, Guy Cox wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> There's no simple answer.  Some people have aimed to use deconvolution to
>> go beyond the optical resolution limit.  The best example I know is:
>> Carrington, W.A., Lynch, R.M., Moore, E.D.W., Isenberg, G., Fogarty, K.E.
>> and Fay, F.S., 1995.  Superresolution Three-Dimensional Images of
>> Fluorescence in Cells with Minimal Light Exposure. Science 268, 1483-1487
>>
>> Commercial deconvolution systems are not usually aiming at that.  The aim
>> is to separate in focus and out of focus light and thereby achieve
>> optical sectioning in wide field.  Since the light budget is always
>> better in wide field than in confocal this has certain advantages, but it
>> has the disadvantage that the wide field OTF has a 'missing cone' of
>> directions in which no information is present.  This limits what can be
>> achieved.
>>
>> The confocal OTF has no missing cone which makes it quite a good target
>> for deconvolution.  The confocal PSF Is far from optimal, being very
>> streaked out in the Z direction, and deconvolution can do a lot to
>> improve it.  Several people (including me) have advocated 1D
>> deconvolution in just the Z direction, which can greatly improve image
>> quality.  My papers on this are a bit hard to find but there is an
>> example in my chapter in Jim Pawley's book.  However deconvolution of
>> confocal images hasn't really caught on as much as it should have.
>>
>>
>> Guy
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]]
>> On Behalf Of Christophe Leterrier
>> Sent: Wednesday, 24 October 2012 3:28 AM
>> To: [hidden email]
>> Subject: Question about deconvolution
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi folks,
>>
>> I have a long-standing question regarding deconvolution (as processing
>> widefield or confocal images to reassign light from where it originated
>> using a PSF).
>>
>> Is there a theoretical limit to the resolution one could obtain using
>> deconvolution? Is is theoretically possible to "break" the diffraction
>> limit with deconvolution? That is, to get under the classical
>> 200x200x600nm spot? I think it is not the case, but then why would you
>> deconvolve widefield or confocal images? What do you gain by doing so on
>> a system that is reasonably close to its theoretical capabilities in
>> terms of optical performances?
>>
>> Thanks for your help,
>>
>> Christophe
>>
>> --
>> Christophe Leterrier
>> Researcher
>> Axonal Domains Architecture Team
>> CRN2M CNRS UMR 7286
>> Aix Marseille University, France
>>
>>