Re: Problem with 'penetrating' signals in the 2nd channel

Posted by Ke Peng on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Problem-with-penetrating-signals-in-the-2nd-channel-tp7579271p7579274.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Thanks a lot for your reply. In our system excitation are provided by laser
lines: 405 (Diode) and 561 (DPSS). So maybe the excitation filter is not
relevant in this case?

Aromis

-----邮件原件-----
发件人: Confocal Microscopy List [mailto:[hidden email]]
代表 Ryan, Kevin
发送时间: Montag, 5. November 2012 22:13
收件人: [hidden email]
主题: Re: Problem with 'penetrating' signals in the 2nd channel

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

This sounds like a cross-talk issue. Check your excitation filters - since
the emission filter you're using has a dual-band acceptance with broad
bandpasses, only your excitation filter determines what fluoresces. See if
the 405 excitation filter you're using overlaps the Alexa 568 excitation
curve at all. I expect that it does - and it doesn't take much for visible
crosstalk.

Options:

- Use a different set of excitation filters to minimize crosstalk.

- Correct your data afterwards: determine the percentage of the Alexa 568
crosstalk using an Alexa 568 only calibration sample (note any exposure time
- I suggest calibrating with equal times), then post-process to subtract the
Alexa 568 image(s) times that scaling from the Alexa 405 image(s).

        405_image est. = 405_image - [ 568_image * (405 time)/(568 time) *
568_crosstalk_scalar ]

       
Kevin Ryan
Media Cybernetics, Inc.


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Ke Peng
Sent: Monday, November 05, 2012 3:42 PM
To: [hidden email]
Subject: Problem with 'penetrating' signals in the 2nd channel

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear Everyone:

 

I'm doing multi-color sequential imaging of a protein complex and have met a
confusing situation:

 

I used Alexa 568 labeled 2nd Ab to probe one protein and I found that the
signal from the Alexa 568 could be detected in the 405 channel, the channel
that is supposed to detect dyes like Alexa 405. We are using the Pelkin
Elmer spinning disc microscope and the emission filter used is filter 3
(445nm, 615nm). This filter is used for emission discrimination for both
Alexa 405 and Alexa 568. The sequence I used in the imaging was 1: Alexa
568; 2: eGFP; 3: Alexa 405. The excitation and emission wavelengths of these
dyes are well separated and should work well in multicolor imaging.

 

I was wondering whether this is due to the fact that the large amount of
Alexa 568 excited in the first step kept emitting and photons were captured
again through the filter (445nm, 615nm) while signals were being detected
for the Alexa 405 in the 3rd step. It does not sound very likely as the
fluorophore should 'finish' emitting within the range of nanoseconds when
excitation is removed and there is an eGFP detection (lasts about 300ms)
step in between the two detection. But I have no other explanation at the
moment. Does anyone of you have similar experience and/or have a clue how
this happened?

 

Thanks a lot for your help in advance.

 

Best regards,

 

Aromis
############################################################################
##########
CONFIDENTIALITY NOTICE:
This email transmission and its attachments contain confidential and
proprietary information of Princeton Instruments, Acton Research, Media
Cybernetics and their affiliates and is intended for the exclusive and
confidential use of the intended recipient. Any use, dissemination,
printing, or copying of this transmission and its attachment(s) is strictly
prohibited. If you are not the intended recipient, please do not read,
print, copy, distribute or take action in reliance upon this message.  If
you have received this in error, please notify the sender immediately by
telephone or return email and promptly delete all copies of the original
transmission and its attachments from your computer system.
############################################################################
###########