Re: formula for z-resolution

Posted by Arne Seitz on
URL: http://confocal-microscopy-list.275.s1.nabble.com/AQLM-2013-Last-chance-tp7579522p7579524.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear Steffen,

I found (and use) the following definition for z-resolution (Jerome Mertz, Optical Microscopy, Roberts and Company Publishers, p. 297)

delta z (WF)=n*lambda)/NA^2
 
delta z(conf)=n*lambda/2*NA^2

Both formulas are derived from the frequency support of the different microscopes. Thus theoretically the frequency of a confocal microscope is twice as good as that of a wide-field microscope. This is (and the author is explicitly stating it) overly optimistic and only true in the case of an infinitesimal small pinhole. Thus is will be hardly reached in practice. Nevertheless I like these definitions as the can be obtained from "pure" theory.

The formula I presented here is identical with your second one. To my understanding his is often named "depth of field" due to the missing cone problem.

Just my 2 c.

Regards
Arne

   

> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:[hidden email]] On Behalf Of Steffen
> Dietzel
> Sent: lundi 21 janvier 2013 09:01
> To: [hidden email]
> Subject: formula for z-resolution
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear confocalists,
>
> I am confused about the correct formula for diffraction limmited resolution
> along the z-axis. Starting with conventional fluoresence
> microscopy:
>
> I used to use the following formula given by Inoue in the first chapter of the
> Handbook
>
> (1) z-min = 2*lambda*n /NA^2
>
> where lambda is the wavelength in air, n the refraction index of the
> immersion medium, NA the numerical aperture of the objective and ^2
> means to the power of 2.
> The text says that this is the distance from the center of the peak to the first
> minimum of the diffraction pattern.
> The same is said by F Quercioli in Diaspro's "Optical Fluorescence
> microscopy".
>
>
>
> In the new Murphy and Davidson (Fundamentals of Light Microscopy and
> Electronic Imaging, 2nd edition, page 109) I find the following formula:
>
> (2) z = lambda*n /NA^2
>
> Note that the "2" is missing, suggesting a resolution twice as good.
> However, this is not explained as Rayleigh criterion but as "depth of field"
>
>
>
> Formula (2) is also given as "resolution in a conventional microscope"
> defined as "distance between points where the intensity is 80% of the peak
> intensity"  by Amos, McConnell and Wilson (Confocal Microscop, Chapter in
> Handbook of Comprehensive Biophysics), but only for cases with an NA <0.5.
> (Note that the clasical Rayleigh criterion in the focal plane leads to 73,5 %
> intensity at the minimum between peaks)
>
> For high NA objectives Amos et al give the following Depth of field = 80%
> limit:
>
> (3) 0.51*lambda/(n-sqrt(n^2-NA^2))
>
> This paper also gives a formula for theoretical confocal/two photon, although
> not for resolution but for FWHM, so that is a little different.
>
>
> Example: 500 nm, NA=1.4, n =1.515, resolution according to the various
> formulas:
>
> (1) 773 nm
> (2) 386 nm
> (3) 272 nm
>
> This sounds very wrong and my gut feeling is I missed something. I'd be
> happy if you could clarify this for me.
>
> Steffen
> --
> ------------------------------------------------------------