Re: BTU generation by microscope components

Posted by Chris Tully on
URL: http://confocal-microscopy-list.275.s1.nabble.com/BTU-generation-by-microscope-components-tp7579617p7579619.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Michael,

Electrical components have two critical characteristics. The Maximum Power
is the the maximum the instrument should ever draw, usually at start up or
under extreme load. Since most devices are designed to operate at somewhere
around 80% of maximum, to calculate the cooling necessary for a given set
of instruments you are better off measuring the average power draw over a
day or two of normal usage. The manufacturers should be able to supply this
number if it is not on the instruments info plate or in the manual.  If not
look for a "Kill A Watt" - this is a device that plugs into the wall and
has a plug for what ever device you wish to measure, and measures the
actual power drawn.

Keep in mind though that the heat you are concerned about is waste heat,
the more efficient a given device is the less waste heat it will produce,
and the more useful work you will get out of each watt consumed.

In all honesty, rather than trying to do the calculations myself, I would
call a few local commercial HVAC contractors and talk with them.  Anyone
who has done the cooling for a large server rrom should be able to give you
far better data than you can come up with on your own.

Chris Tully, MS Materials Science and Engineering
Microscopy and Image Analysis Expert
[hidden email]
240-475-9753 (c)

[image: View my profile on LinkedIn]<http://www.linkedin.com/in/christully/>


On Wed, Jan 30, 2013 at 9:49 AM, Cammer, Michael <[hidden email]
> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi microscopists,
>
> Back in 2011 there was a discussion here about BTUs generated by a
> microscope (
> http://lists.umn.edu/cgi-bin/wa?A2=ind1107&L=CONFOCALMICROSCOPY&P=R5803&D=0) but I can't find an answer in the archives.
>
> The question is, how many BTUs does a microscope system generate.  To be
> more specific, how many BTUs does each of these components generate?
> 1.  Forced air environmental chamber (Plexiglas box by Zeiss or
> Weatherstation) at 37 degrees C.
> 2.  EXFO lamp.
> 3.  AR laser (Lasos).
> 4.  Two computers and 24 inch screen.
> 5.  Scan box (Zeiss 710).
> 6.  MaiTai laser operating at 12 W.
> 7.  Two people (let's say 75 W each).
>
> Does anybody have calculations for this?
>
> Back in the mid 1990s I summed the maximum power draw into each microscope
> room and converted to BTUs, but this resulted in excessive cooling.  Does
> anyone have a source for actual measurements?
>
> Thank you!!
>
> ________________________________________________________
> Michael Cammer, Assistant Research Scientist
> Skirball Institute of Biomolecular Medicine
> Lab: (212) 263-3208  Cell: (914) 309-3270
>