Martin,
I must say I am very worried by your statement "Antibodies that work for one application (e.g., westerns) may be completely unusable for another (e.g., IF)". Can you give me a biological reason for this? It would seem to make most cell biological work impossible. I would assume that most of us work the same way as me - ie using microscopy to identify the location, and chromatography to characterise the protein. If you don't use the same antibody both times you will just be generating nonsense!
Guy
________________________________________
From: Confocal Microscopy List [
[hidden email]] on behalf of Martin Wessendorf [
[hidden email]]
Sent: 09 March 2013 02:31
To:
[hidden email]
Subject: Re: antibodies for immunofluorescence
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Dear Michal--
On 3/8/2013 2:36 AM, MichaĆ Majkowski wrote:
> I do not know if this is a good place to address my problem as it
> regards immunofluorescence. The problem is: we have in the lab
> antibodies dedicated to Western Blot (according to manufacturer; Abcam).
> We used them to IF and we obtained some nice images. What do you think
> about such data? Should it be confirmed by IgG dedicated to IF? Thank
> you for help.
The quick and safe answer is that you can never trust any antibody for
any purpose without first characterizing it thoroughly.
Antibodies that work for one application (e.g., westerns) may be
completely unusable for another (e.g., IF)...or they may work perfectly.
A starting point is to compare the labeling that you obtain, to the
labeling that's previously been published. The "gold standard" is to
demonstrate that the labeling is not observed in a knockout animal.
Cliff Saper at Journal of Comparative Neurology published an editorial
on antibody use that may be helpful. It can be found at:
http://www.wiley.com/legacy/wileyblackwell/images/antibody_editorial .
Good luck!
Martin Wessendorf
--