Hi Guy,

I don't think what Martin said is at all controversial.  Some antibodies are very good at recognizing epitopes that are unmasked only after proteins have been denatured by SDS-PAGE, making them useful for western blotting.  Other antibodies can recognize epitopes that are present in the native protein, making them useful for IF, but epitopes can be destroyed by denaturing, which makes antibodies that specifically recognize them bad for westerns.  As long as you have appropriate controls to show the specificity for each antibody in each assay, I see no reason why one couldn't use two different antibodies for the same protein, depending upon the application.

Eric


Eric Griffis
Principal Investigator
Centre for Gene Regulation and Expression
College of Life Sciences
University of Dundee
MSI/WTB/JBC Complex
Dow Street
Dundee DD1 5EH
UK

Tel: +44 (0)1382 385118
Fax: +44 (0)1382 388072
Email: [hidden email]<mailto:[hidden email]>
Lab Website:  http://www.lifesci.dundee.ac.uk/groups/eric_griffis/GriffisLab/index.html
GRE Website: http://gre.lifesci.dundee.ac.uk



On 8 Mar 2013, at 16:02, Guy Cox wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Martin,

             I must say I am very worried by your statement "Antibodies that work for one application (e.g., westerns) may be completely unusable for another (e.g., IF)".  Can you give me a biological reason for this?  It would seem to make most cell biological work impossible.  I would assume that most of us work the same way as me - ie using microscopy to identify the location, and chromatography to characterise the protein.  If you don't use the same antibody both times you will just be generating nonsense!

                                                                             Guy

________________________________________
From: Confocal Microscopy List [[hidden email]] on behalf of Martin Wessendorf [[hidden email]]
Sent: 09 March 2013 02:31
To: [hidden email]
Subject: Re: antibodies for immunofluorescence

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear Michal--

On 3/8/2013 2:36 AM, MichaƂ Majkowski wrote:

I do not know if this is a good place to address my problem as it
regards immunofluorescence. The problem is: we have in the lab
antibodies dedicated to Western Blot (according to manufacturer; Abcam).
We used them to IF and we obtained some nice images. What do you think
about such data? Should it be confirmed by IgG dedicated to IF? Thank
you for help.

The quick and safe answer is that you can never trust any antibody for
any purpose without first characterizing it thoroughly.

Antibodies that work for one application (e.g., westerns) may be
completely unusable for another (e.g., IF)...or they may work perfectly.
 A starting point is to compare the labeling that you obtain, to the
labeling that's previously been published.  The "gold standard" is to
demonstrate that the labeling is not observed in a knockout animal.

Cliff Saper at Journal of Comparative Neurology published an editorial
on antibody use that may be helpful.  It can be found at:
http://www.wiley.com/legacy/wileyblackwell/images/antibody_editorial .

Good luck!

Martin Wessendorf

--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455                    e-mail: [hidden email]


The University of Dundee is a registered Scottish Charity, No: SC015096