Re: colocalization between confocal and two-photon: voxel matching
Posted by
Jeremy Adler-4 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/xyzt-image-tp7580102p7580115.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Surprisingly actually measuring colocalization is surprisingly tricky
- there are far too many coefficients and nonsense to several decimal
places is still nonsense.
It is also important to appreciate that colocalization addresses two
separate questions (1) cooccurence, are the molecules in the same
pixels and (2) correlation, is there any relationship between the
intensities. These cannot be measured with a single coefficient.
It is also important to appreciate that the quality, meaning noise, of
the individual images seriously alters the measured colocalization.
Try taking a second image and comparing it to the first - they are
unlikely to be identical and for a correlation measurement this noise
creates a difference between the measured and the true (noise free)
correlation. A solution to the problem of noise is available for
correlation measurements.
look at
http://diva-portal.org/smash/record.jsf?searchId=11&pid=diva2:563664Jeremy Adler