Hi Claire,
Are these on inverted stands? At least up to the Axio line, Zeiss lenses telescoped with the nosepiece retracting into the barrel. Oil can seep down through the seal where the nose retract into the barrel. A "scrunchie" hair retainer wrapped around the nosepiece will retard the drips and provide time to check lenses periodically.
My problem is motorized stages, not oil. In a collision, the stage always wins. My current quote from Zeiss to repair a 100X oil is about US$3300.
Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Cellular Morphology Core
Center on Human Development and Disability
Box 357923
University of Washington
Seattle, WA 98195-7923 USA
(206) 616-4156
[hidden email]
On May 27, 2013, at 4:37 PM, "Claire Brown, Dr." <
[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> We are relatively routinely seeing oil inside of our Zeiss 63x oil immersion lenses. You end up with a bead of oil inside that acts as a lens itself and the lens is unusable.
>
> As far as we can tell, it usually happens after prolonged use over many years. We are at a point now where our lenses are "old" and Zeiss cannot repair them because the parts are no longer available. It costs us $6-8,000 to replace them. With 13 microscopes we have one lens failing every 6-12 months or so. I can't afford to keep replacing these lenses and with this becoming a "routine" occurrence I want to seek out advice about why this might be happening. Our users are well trained and I don't think it is neglect by them. I think this is due to routine use.
>
> I have two questions:
>
>
> 1) "Is there a defect in how these lenses are made?"
>
> For example, perhaps the seals on the front lenses degrade over time - we inspect our lenses regularly and many that have had oil inside have no obvious damage by visual inspection. Maybe they need to be resealed every year or two, maybe an internal seal degrades over time? Preventative maintenance of these lenses or a change in how they are manufactured would certainly cost less than $8,000 I would think.
>
>
>
> 2) "Do people have a similar problem with oil immersion lenses from the other major manufacturers?"
>
> I look forward to some input here.
>
> Sincerely,
>
> Claire
>
> -------------------------------------------------------------------------------------------
> CANADIAN CYTOMETRY AND MICROSCOPY SYMPOSIUM - MONTREAL JUNE 13-15 -
http://www.regonline.com/ccma-accm2013>
> Claire M. Brown, PhD - McGill University - Assistant Professor, Physiology - Life Sciences Complex Imaging Facility Director