Re: autofluorescence in liver tissue

Posted by Kilgore, Jason-2 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/autofluorescence-in-liver-tissue-tp7580412p7580417.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Autofluorescence can be especially bad for paraffin-embedded sections.  I'm not sure liver is particularly worse, though, compared to many tissues.

I've had very good luck with sodium borohydride washing to reduce autofluorescence, particularly in the green and red range.

Another thing I've found to reduce it is to use Histoclear or other limonine solvents for deparaffinization, instead of xylenes.

Another thing to try is to amplify your label, to help overcome the background, using biotin/streptavidin, Tyramide signal amplification, or Qdots.

Hope that helps.  Cheers,

Jason


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Cromey, Douglas W - (dcromey)
Sent: Tuesday, June 04, 2013 10:24 AM
To: [hidden email]
Subject: autofluorescence in liver tissue

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I am working with a lab looking at IF stained sections of formalin-fixed, paraffin embedded pieces of mouse or rat liver.  If there was any actual staining in their first try on our confocal, it was largely washed out by the lovely and strong autofluorescence.  Because their animal experiment takes 8 weeks, it's not trivial (or inexpensive) to get new tissue, so they have been using existing blocks.  No one in their lab had the foresight to put up snap-frozen tissues for cryosectioning.  I have them looking at this document: http://www.uhnres.utoronto.ca/facilities/wcif/PDF/Autofluorescence.pdf but I'm wondering if liver is always a problem with autofluorescence, or is there maybe something in this lab's sample prep protocol that's making it worse.  Any ideas?

Doug

^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
Douglas W. Cromey, M.S. - Associate Scientific Investigator
Dept. of Cellular & Molecular Medicine, University of Arizona
1501 N. Campbell Ave, Tucson, AZ  85724-5044 USA

office:  AHSC 4212         email: [hidden email]
voice:  520-626-2824       fax:  520-626-2097

http://swehsc.pharmacy.arizona.edu/micro
Home of: "Microscopy and Imaging Resources on the WWW"