Eventually, the quenchers in the media will be internalized in live cells, simply due to repeated endocytotic uptake. Usually you have at least 45-60 minutes before this is a problem, but it depends on the rate of endocytosis of your cell line.
If you cells die or in some other way have the membranes compromised, then certainly the external quencher will now quench internally as well. As one other commenter pointed out, this isn't an issue if you are only wanting to image live cells, though.
Thank you for all valuable inputs. We can certainly try to quench the
fluorophore on the cell surface. However, I am wondering if the quencher
or incubated with the quenchers for too long. Some treatment seem quite
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> I can confirm Michaels comment below. During my research life, I was
> labelling bacteria with FITC, and observing uptake by cultured macrophages.
> Due to the conditions used, a lot of these FITC-bugs stuck to the outside
> of the cell, and I used Typan blue to quench the extracellular
> fluorescence. Details in
http://www.ncbi.nlm.nih.gov/pubmed/11726395>
> TITLE: Effects of surfactant protein A and NaCl concentration on the
> uptake of Pseudomonas aeruginosa by THP-1 cells. Khubchandani-Aswani KR,
> Oberley RE, Snyder JM.
>
> Cheers,
> Kavita
>
> Kavita Aswani, PhD
> Senior Applications Scientist - Life Sciences
> Lumen Dynamics Group Inc.
> 2260 Argentia Road, Mississauga, Ontario, Canada, L5N 6H7
> Direct Tel. 905 812-3342| Cell. 647 290 3506 * Toll Free (USA and Canada) 1
> 800 668-8752
>
[hidden email] *
http://www.ldgi.com/>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:
[hidden email]]
> On Behalf Of MODEL, MICHAEL
> Sent: June-07-13 7:04 AM
> To:
[hidden email]
> Subject: Re: How to show something inside the cell?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> I think trypan blue has also been used to quench external fluorescence
> ________________________________________
> From: Confocal Microscopy List [
[hidden email]] on
> behalf of Gabriel Lapointe [
[hidden email]]
> Sent: Thursday, June 06, 2013 5:17 PM
> To:
[hidden email]
> Subject: Re: How to show something inside the cell?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Hi Yi,
> In the same line of thougth as Mike, you could also look at the quencher
> used in qPCR. Some might be compatible with your fluorophore.
>
>
>
> *Gabriel Lapointe, M.Sc.*
> Lab Manager / Microscopy Specialist
> Concordia University, Biology Department
> 7141 Sherbrooke St. West SP 534
> Montréal QC H4B 1R6 Canada
> Lab : (514) 848-2424 x5988
> Office : (514) 848-2424 x3008
> Fax : (514) 848-2881
> Cell : (514) 278-0247
>
[hidden email]
> cmac.concordia.ca
>
http://gabriellapointe.ca>
>
> 2013/6/6 MODEL, MICHAEL <
[hidden email]>
>
> > Hi Yi - I think your solution is good. Fluorophores on the surface can
> > also be selectively quenched, depending on the fluorophore - FITC by
> > lowering pH, red fluorophores or those emitting in the 600s by adding
> > several mg/ml of Acid Blue 9.
> >
> > Mike Model
> >
> >
> > -----Original Message-----
> > From: Confocal Microscopy List
> > [mailto:
[hidden email]]
> > On Behalf Of SUBSCRIBE CONFOCALMICROSCOPY
> > Sent: Thursday, June 06, 2013 4:27 PM
> > To:
[hidden email]
> > Subject: How to show something inside the cell?
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > *****
> >
> > Dear listers,
> >
> > One of our user is investigating the transportation of fluorescence
> > labeled DNA oligo in cells. He would like to prove the DNA oligo has
> > been transported inside the cells but not on the surface of cells by
> > using the light microscopy. We suggested him to clearly label the
> > membrane of cells and use the Z-stack images of confocal to look through
> the cells.
> > I am wondering if there is any other better solution?
> >
> > Any suggestion on good membrane dye is appreciated as well!
> >
> > Thank you,
> >
> > Yi
> >
> >
> >
>
> Think Green Before You Print!
>
> This email and any attachments thereto may contain private, confidential,
> and privileged material for the sole use of the intended recipient. Any
> review, copying, or distribution of this email (or any attachments thereto)
> by others is strictly prohibited. If you are not the intended recipient,
> please contact the sender immediately, and permanently delete the original
> and any copies of this email and any attachments thereto.
>