Almost none, the Sutter uses common narrow band interference filters that are well blocked outside the pass band. As you raised this question, I wonder how much heat comes from a LED source -the chip gets very hot in high power applications and about half the power consumption of a LED is waste heat…

Cheers M

 
On 6/11/2013, at 1:48 am, Cammer, Michael <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> How much of the 70 mW is from UV or IR leaking through?
> _________________________________________
> Michael Cammer, Assistant Research Scientist
> Skirball Institute of Biomolecular Medicine
> Lab: (212) 263-3208  Cell: (914) 309-3270
>
> ________________________________________
> From: Confocal Microscopy List [[hidden email]] on behalf of Julio Vazquez [[hidden email]]
> Sent: Tuesday, November 05, 2013 5:09 PM
> To: [hidden email]
> Subject: Re: Brightness difference Hg vs LED
>
>
> A quick comparison in our lab gave the following:
>
> 488 nm (FITC) band of an LED-based light engine (Lumencor): 40 mW
> 300 W Xenon Sutter illuminator (probably our brightest light source) with FITC EX filter: 70 mW
>
> both measured at the exit of the fiber or light guide.
>
> I agree that LED based light engines are probably at least as good (power-wise) as most lamps, plus they have other benefits. Charts such as the one found here should be a good guide regarding relative power:
>
> http://lumencor.com/wp-content/uploads/2011/09/SpectraX-LE-User-Manual1.pdf
>
>
> Julio Vazquez, PhD
> Fred Hutchinson Cancer Research Center
> Seattle, WA 98109
>
> http://www.fhcrc.org/en.html
>
>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
> =================================

Mark  B. Cannell Ph.D. FRSNZ