Hi All,

I’m sure there is going to be a simple answer here, but it alludes me.

I know the psf is a function of NA and wavelength.

Obviously, with 1P we use a pinhole to exclude out of focus light, and when set at ā€œ1 Airy,ā€ you have maximized the pinhole for the particular wavelength you are using.  With 2P, no pinhole is necessary because of the non-linear excitation mechanism.

A user just approached me and asked if/why there would be more out of focus light in a 2P image then in a 1P image with a properly set pinhole.  In a collaborators experiments, there seems to be more background in the 2P image (all other things equal).

Only thing I can think of is a poor beam profile on the 2P.  Maybe a large pulse width would excite a larger spot?

Thankful for the insight.


Best,

Gary