Hi Gary,

Is your collaborator using a non-descanned detector?  A computer monitor
near the scope could produce enough background light to interfere with his
imaging.  Those pick up stray light like a widefield scope does, sometimes
worse.  

All the best,


TF

Timothy Feinstein, Ph.D. | Confocal Manager
333 Bostwick Ave., N.E., Grand Rapids, Michigan 49503
Phone: 616-234-5819 | Email: [hidden email]







On 11/15/13, 9:00 AM, "Laevsky, Gary S." <[hidden email]> wrote:

>*****
>To join, leave or search the confocal microscopy listserv, go to:
>http://scanmail.trustwave.com/?c=129&d=rqmG0jAq5wSLcDwzveugGEfcTBlOgb6VWnC
>SP5ImQw&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfocalmic
>roscopy
>*****
>
>Hi All,
>
>I¹m sure there is going to be a simple answer here, but it alludes me.
>
>I know the psf is a function of NA and wavelength.
>
>Obviously, with 1P we use a pinhole to exclude out of focus light, and
>when set at ³1 Airy,² you have maximized the pinhole for the particular
>wavelength you are using.  With 2P, no pinhole is necessary because of
>the non-linear excitation mechanism.
>
>A user just approached me and asked if/why there would be more out of
>focus light in a 2P image then in a 1P image with a properly set pinhole.
> In a collaborators experiments, there seems to be more background in the
>2P image (all other things equal).
>
>Only thing I can think of is a poor beam profile on the 2P.  Maybe a
>large pulse width would excite a larger spot?
>
>Thankful for the insight.
>
>
>Best,
>
>Gary
>
>
>