The 2P may cause more background due to the broad excitation spectra exciting other molecules in the specimen -i.e. effectively more non specific label shows up.

Cheers Mark

On 15/11/2013, at 2:00 pm, Laevsky, Gary S. <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi All,
>
> I’m sure there is going to be a simple answer here, but it alludes me.
>
> I know the psf is a function of NA and wavelength.
>
> Obviously, with 1P we use a pinhole to exclude out of focus light, and when set at ā€œ1 Airy,ā€ you have maximized the pinhole for the particular wavelength you are using.  With 2P, no pinhole is necessary because of the non-linear excitation mechanism.
>
> A user just approached me and asked if/why there would be more out of focus light in a 2P image then in a 1P image with a properly set pinhole.  In a collaborators experiments, there seems to be more background in the 2P image (all other things equal).
>
> Only thing I can think of is a poor beam profile on the 2P.  Maybe a large pulse width would excite a larger spot?
>
> Thankful for the insight.
>
>
> Best,
>
> Gary
>
>
>
> Gary Laevsky, Ph.D.
> Confocal Imaging Facility Manager
> Dept. of Molecular Biology
> Washington Rd.
> Princeton University
> Princeton, New Jersey, 08544-1014
> (O) 609 258 5432
> (C) 508 507 1310

Mark  B. Cannell Ph.D. FRSNZ
Professor of Cardiac Cell Biology
School of Physiology &  Pharmacology
Medical Sciences Building
University of Bristol
Bristol
BS8 1TD UK

[hidden email]