Don't forget to check the power requirements of the equipment your going to
plug in those UPS. If they are under powered they may cause more problem
than they solve. I installed our swepfield on two 1,2KW UPS to spread the
load. Leaving me a few 100s watts to spare, but not that much. They waren't
cheap, but neither are the systems they are meant to protect.
*Gabriel Lapointe, M.Sc.*
Lab Manager / Microscopy Specialist
Concordia University, Biology Department
7141 Sherbrooke St. West SP 534
Montréal QC H4B 1R6 Canada
Lab SP-534: (514) 848-2424 x5988
Office SP-553.05: (514) 848-2424 x3008
Fax : (514) 848-2881
Cell : (514) 278-0247
[hidden email]
cmac.concordia.ca
http://gabriellapointe.ca2013/11/27 Craig Brideau <
[hidden email]>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Cyberpower or APC are usually pretty good. Just remember that you will
> have to replace the batteries every 3-4 years approximately.
>
> Craig
>
>
> On Wed, Nov 27, 2013 at 1:30 PM, Vladimir Ghukasyan
> <
[hidden email]>wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > *****
> >
> > Hello All,
> >
> > We have seen a few problems in the facility caused by power problems. I'm
> > looking for a good and not very expensive UPS to help with that and would
> > be thankful if you could share what are you using and whether you're
> happy
> > with them.
> >
> > Thank you in advance and Happy Thanksgiving.
> >
> > Bests,
> > Vladimir
> >
>