Michael,
The 40X/1.4 is a good objective for dim samples on a widefield system, I
have used it quite a lot.
The 40X/1.4 has:
24.0 brightness
161nm sampling on a 6.5µm pixel (ie Flash 4.0 CMOS)
333x333µm field of view on the Flash4.0
The 63X/1.4 has:
9.7 brightness
102nm sampling on a 6.5µm pixel (ie Flash 4.0 CMOS)
211x211µm field of view on the Flash4.0
If you are going to deconvolve the image you will find you are
undersampling quite a bit with the 40X. When you look through the eyepiece
you will really appreciate the 2.5X improvement in brightness.
Another note is that the 1.4NA 40X and 63X have ~80% transmission from
450-750nm and the new 100X/1.4 from Zeiss has ~90%.
Also, if you are using a spinning disk you will see your axial resolution
suffer with the 40X at least with the CSU-X1 pinholes.
-Ryan Schreiner
Weill Cornell Medical College
On Fri, Jul 11, 2014 at 9:53 AM, Cammer, Michael <
[hidden email]
> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Does anyone have experience with the new Zeiss 40X N.A. 1.4 PlanApo? This
> is something I've wanted for a long time, the ability to take large fields
> of view (2k X 2k pixels) at high resolution instead of having to do tiling.
> Also, with the new cameras that have oodles of small pixels...
>
> I'm considering replacing our 63X with this new 40X. Any experience with
> this, other than the battle of having to explain to other scope users why
> this is not really lower magnification?
>
> Regards,
>
> Michael
>
> ===========================================================================
> Michael Cammer, Microscopy Core & Dustin Lab , Skirball Institute, NYU
> Langone Medical Center
> Cell: 914-309-3270 Lab: 212-263-3208
>
http://ocs.med.nyu.edu/microscopy &
>
http://www.med.nyu.edu/skirball-lab/dustinlab/>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of the
> intended recipient(s) and may contain information that is proprietary,
> confidential, and exempt from disclosure under applicable law. Any
> unauthorized review, use, disclosure, or distribution is prohibited. If you
> have received this email in error please notify the sender by return email
> and delete the original message. Please note, the recipient should check
> this email and any attachments for the presence of viruses. The
> organization accepts no liability for any damage caused by any virus
> transmitted by this email.
> =================================
>