Re: Zeiss 40X N.A. 1.4 Plan APo as replacement for 63X?

Posted by Miroslav Varecha on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Zeiss-40X-N-A-1-4-Plan-APo-as-replacement-for-63X-tp7582324p7582337.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

True, but I was speaking about my LSM 700 and we have only 63x at
NA1,4. We have only NA 1,3 for 40x. It would be great to have NA 1,4
as 40x is most used lens for us.

Miroslav

------------------------------------------------
Miroslav Varecha, Ph.D.
Department of Biology, Faculty of Medicine
Masaryk University, Brno, Czech Republic

2014-07-14 11:13 GMT+02:00 Mark Cannell <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Re: “ But if you want to observe really up close organels and such I would go to 63x NA1,4” .
>
> The beauty of the scanning confocal is that you can use zoom to go from 40x -> 63x or whatever. You do not need an objective of the same NA but different magnification to achieve this.
>
> Cheers,
> Mark
>
>
> On 14/07/2014, at 8:54 am, Miroslav Varecha <[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Hello,
>> I would like to react to George. In my humble opinion, best lateral
>> optical resolution for confocal microscope is around 130 nm, so going
>> to resolution 60x60 nm is overshoot and it is just wasting your drive
>> space as you are not collecting any new real information. Resolutions
>> of 50x50nm and such are area of superresolution microscopy. Our
>> confocal Zeiss LSM 700 we have also highest resolution of 2k x 2k.
>> High quality CCDs have even less, but sCMOS can get higher than 2k x
>> 2k for sure. We have objective Zeiss 40x but NA 1,3 and I can tell you
>> that it is the most used objective (usually we observe stem cells)
>> even tho it is not NA1,4. We have also 63x1,4 NA but ppl prefer larger
>> field of view of 40x. You still get great detail and many cells in one
>> image. But if you want to observe really up close organels and such I
>> would go to 63x NA1,4. 40x NA 1,4 seems to me like most flexible
>> objective for vast majority of observations.
>> Miroslav
>>
>>
>> 2014-07-14 1:31 GMT+02:00 George McNamara <[hidden email]>:
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> Post images on http://www.imgur.com and include the link in your posting.
>>> *****
>>>
>>> Hi Michael,
>>> why 2kx2k?
>>> If the 40x lens has a 250x250 um field of view, this would be undersampling,
>>> pixel size 125x125 nm. If even larger field of view, undersampling even
>>> more.
>>> I suggest pixel size of 50x50 or 60x60 nm, and 3D deconvolution (Z step 200
>>> nm, maybe closer),
>>> George
>>>
>>> On 7/11/2014 8:53 AM, Cammer, Michael wrote:
>>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>> Post images on http://www.imgur.com and include the link in your posting.
>>>> *****
>>>>
>>>> Does anyone have experience with the new Zeiss 40X N.A. 1.4 PlanApo?  This
>>>> is something I've wanted for a long time, the ability to take large fields
>>>> of view (2k X 2k pixels) at high resolution instead of having to do tiling.
>>>> Also, with the new cameras that have oodles of small pixels...
>>>>
>>>> I'm considering replacing our 63X with this new 40X.  Any experience with
>>>> this, other than the battle of having to explain to other scope users why
>>>> this is not really lower magnification?
>>>>
>>>> Regards,
>>>>
>>>> Michael
>>>>
>>>>
>>>> ===========================================================================
>>>> Michael Cammer, Microscopy Core&  Dustin Lab , Skirball Institute, NYU
>>>> Langone Medical Center
>>>> Cell:  914-309-3270   Lab: 212-263-3208
>>>> http://ocs.med.nyu.edu/microscopy&
>>>> http://www.med.nyu.edu/skirball-lab/dustinlab/
>>>>
>>>> ------------------------------------------------------------
>>>> This email message, including any attachments, is for the sole use of the
>>>> intended recipient(s) and may contain information that is proprietary,
>>>> confidential, and exempt from disclosure under applicable law. Any
>>>> unauthorized review, use, disclosure, or distribution is prohibited. If you
>>>> have received this email in error please notify the sender by return email
>>>> and delete the original message. Please note, the recipient should check
>>>> this email and any attachments for the presence of viruses. The organization
>>>> accepts no liability for any damage caused by any virus transmitted by this
>>>> email.
>>>> =================================
>>>>
>>>>
>>>
>>>
>>>
>>> --
>>>
>>>
>>>
>>> George McNamara, Ph.D.
>>> Single Cells Analyst
>>> L.J.N. Cooper Lab
>>> University of Texas M.D. Anderson Cancer Center
>>> Houston, TX 77054
>>> Tattletales http://works.bepress.com/gmcnamara/42
>
> Mark  B. Cannell Ph.D. FRSNZ
> Professor of Cardiac Cell Biology
> School of Physiology &  Pharmacology
> Medical Sciences Building
> University of Bristol
> Bristol
> BS8 1TD UK
>
> [hidden email]