Hi Michael,
because I did not see it mentioned in the discussion so far:
To achieve the high aperture and resolution, you have to make sure
- refractive properties of all media in the light path including immersion
and embedding allow.
- overfill the back aperture of the 40x lens which is considerably wider
than that of a 63x or 100x, so collimation must fit.
I propose to compare the system PSF with the different lenses, before you
approach your facilitiy users.
A further potential drawback: Looking down the eyepiece, samples will look
much brighter than with higher-x lenses, but using suitable sampling, you
will easily burn the sample and saturate detectors, trying to achieve the
brightness seen down the eyepiece.
Cheers Jens
http://br.linkedin.com/pub/jens-rietdorf/6/4a3/189/Skype Jens.Rietdorf
On Fri, Jul 11, 2014 at 10:53 AM, Cammer, Michael <
[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Does anyone have experience with the new Zeiss 40X N.A. 1.4 PlanApo? This
> is something I've wanted for a long time, the ability to take large fields
> of view (2k X 2k pixels) at high resolution instead of having to do tiling.
> Also, with the new cameras that have oodles of small pixels...
>
> I'm considering replacing our 63X with this new 40X. Any experience with
> this, other than the battle of having to explain to other scope users why
> this is not really lower magnification?
>
> Regards,
>
> Michael
>
> ===========================================================================
> Michael Cammer, Microscopy Core & Dustin Lab , Skirball Institute, NYU
> Langone Medical Center
> Cell: 914-309-3270 Lab: 212-263-3208
>
http://ocs.med.nyu.edu/microscopy &
>
http://www.med.nyu.edu/skirball-lab/dustinlab/>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of the
> intended recipient(s) and may contain information that is proprietary,
> confidential, and exempt from disclosure under applicable law. Any
> unauthorized review, use, disclosure, or distribution is prohibited. If you
> have received this email in error please notify the sender by return email
> and delete the original message. Please note, the recipient should check
> this email and any attachments for the presence of viruses. The
> organization accepts no liability for any damage caused by any virus
> transmitted by this email.
> =================================
>