Re: Zeiss 40X N.A. 1.4 Plan APo as replacement for 63X?

Posted by Steffen Dietzel on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Zeiss-40X-N-A-1-4-Plan-APo-as-replacement-for-63X-tp7582324p7582353.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Am 14.07.2014 16:28, schrieb Feinstein, Timothy:
> You are correct that even in the best case a traditional optical
> microscope can only separate point sources > 100 nm apart.  

Except of course the two point sources are in different color channels.
Then you can get the two much closer together and still measure the
distance between their intensity centers very accurately. In the
two-digit nm range, depending only on your positioning accuracy.
Provided you correct for chromatic aberration.

Cheers
Steffen


--
------------------------------------------------------------