Thinking about this a bit more, I realised that your existing grid is perfectly suitable as you don’t need 2P excitation to measure field distortion at the wavelength you are using, Just turn the power down a lot and record a reflection image,
>>
>> Hi Mark,
>>
>> I'm assuming you mean beads deposited in a grid or something similar?
>> Could you point me to where you found this?
>>
>> Thanks,
>> Mike
>>
>> On Tue, Jul 22, 2014 at 3:58 AM, Mark Cannell
>> <
[hidden email]> wrote:
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>> Post images on
http://www.imgur.com and include the link in your
>> posting.
>>> *****
>>>
>>> Hi Michael
>>>
>>> I’ve always used beads to test the microscope. They are relatively cheap
>> and if you blow one up it don’t matter. As you say, evaporated metal
>> targets are too easily damaged (as Jim Pauley and I discovered when testing
>> my microscope back in ’95!)
>>>
>>> HTH
>>>
>>> Mark
>>>
>>> On 22/07/2014, at 4:30 am, Michael Giacomelli <
[hidden email]> wrote:
>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>>> Post images on
http://www.imgur.com and include the link in your
>> posting.
>>>> *****
>>>>
>>>> Could someone recommend a good multiphoton distortion target?
>>>> Preferably one that is relatively robust against damage.
>>>>
>>>> Currently I am using a thorlabs wire grid target with fluorophore
>>>> behind it. The beam is attenuated by the metal, resulting in an image
>>>> of the grid. However, passing through the entire 1 mm slide greatly
>>>> reduces my resolution, and unless I am extremely careful, the grid is
>>>> ablated by the beam.
>>>>
>>>> Thanks,
>>>> Michael
>>>
>>>