It is interesting to see this idea go commercial. We computed
transverse and axial-resolution of this scheme (which we call pixel
reassignment): https://www.researchgate.net/publication/254260061_Superresolution_by_image_scanning_microscopy_using_pixel_reassignment?ev=prf_pub
We found that (Fig. 5 of above paper) if the detector array is close
to the size of the airy disk, you actually don't lose too much
sectioning, but gain significant SNR, as also noted in the white
paper.

The original idea of pixel reassignment was proposed by Colin Sheppard in 1988:
https://www.researchgate.net/publication/235994020_Super-resolution_in_confocal_imaging

Shalin
On Fri, Jul 25, 2014 at 8:33 AM, John Oreopoulos
<[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Zdenek,
>
> I think you're right. When I read this white paper again closely, it actually sounds more like what was recently branded "re-scan confocal microscopy":
>
> De Luca, G. M. R., Breedijk, R. M. P., Brandt, R. A. J., Zeelenberg, C. H. C., de Jong, B. E., Timmermans, W., et al. (2013). Re-scan confocal microscopy: Scanning twice for better resolution.Biomed. Opt. Express, 4, 2644-2656.
>
> A lot of the same references are in this article, and this also cites Enderlein and York.
>
> Cheers,
>
> John Oreopoulos
>
>
> On 2014-07-25, at 4:09 AM, Zdenek Svindrych wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>>
>> Very interesting.
>>
>> It's most likely based on the Image Scanning Microscopy (ref 4 of the white
>> paper; it's a physics journal, so the style is a bit different), it allows
>> up to 2x resolution improvement (as all the  other linear structured
>> illumination methods), but with the Enderlain's image processing you
>> actually sacrifice axial resolution if you want high SNR (corresponds to
>> bigger pinhole). Unfortunately they don't provide much details in the white
>> paper about the new image processing approach that gives also axial
>> superresolution...
>>
>> Moreover, I think this approach is not quite compatible with other benefits
>> of modern confocals, such as spectral detection or lifetime imaging.
>>
>> Looking forward to seeing further advances in this direction.
>>
>> Zdenek Svindrych, Charles univ., Czech Rep.
>>
>>
>>
>> ---------- Původní zpráva ----------
>> Od: John Oreopoulos <[hidden email]>
>> Komu: [hidden email]
>> Datum: 25. 7. 2014 4:30:32
>> Předmět: Re: Johannes Amon posted new Zeiss white paper on airyscanning at .
>> .
>>
>> "*****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> The white paper states:
>>
>> "An acentric, shifted pinhole detector produces an image of about the same
>> resolution as a pinhole detector which is aligned to the optical axis,
>> although smaller in amplitude and shifted by half the displacement. This
>> insight has been the motivation for constructing an area detector for a
>> confocal microscope. Such a detector should cover more than 1 AU and contain
>> multiple sub-Airy detector elements. Detection efficiency will be
>> significantly increased by reassigning the detected photons from the shifted
>> detector elements to the central detection position and summing up the back
>> shifted signal from all detector elements. [2] No light is rejected by a
>> closed pinhole but instead collected by the off-axis detector elements.
>> Therefore an increased signal level arises from the reassignment of photons
>> to a smaller spatial region."
>>
>> It sure does sound a lot like Jim Pawley's array detector pinhole idea to
>> me, which is outlined in quite clearly chapter 2 of the Handbook of
>> Biological Confocal Microscopy (3d Edition):
>>
>> http://link.springer.com/chapter/10.1007/978-0-387-45524-2_2
>>
>> See Figure 2.15 in particular. There isn't much detail in the white paper to
>> know what the differences are however, and there is no citation of Pawley's
>> work on this. Here's the relevent passage from the chapter:
>>
>> "In addition to high QE, Si photon detectors have a variety of other
>> practical advantages. As the sensitive element in such a detector is
>> typically very small (5–30 mm on a side), selective use of only a few
>> elements in a small, planar, 2D array could permit it to operate in the CLSM
>> as a combination pinhole and detector. Figure 2.15 is a sketch of what such
>> a detector might look like. After each pixel interval of the microscope, the
>> charge pattern in the 5 x 5 sensor array at the top would be transferred to
>> the read register and then the signal in all 25 pixels would be read out
>> sequentially at about 35MHz. These 25 values could then be “decoded” in a
>> number of possible ways, the most straightforward of which would be to
>> provide three separate signals corresponding to the summed signals from the
>> brown, orange, and red areas of the sensor array. In this way, it would be
>> possible to collect signal simultaneously at three different pinhole sizes.
>> With such a detector, pinhole alignment could be done elec- tronically
>> simply by searching for the detector element producing the most signal from
>> a planar specimen and misalignment could be detected on the fly by
>> comparing, for example, summed output from the 5 pixels on the left with the
>> 5 on the right (Pawley, 1996)."
>>
>> The other reference from 1996 has even more details about this idea:
>>
>> CCDiode: an optimal detector for laser confocal microscopes
>> http://spie.org/Publications/Proceedings/Paper/10.1117/12.237490
>>
>> John Oreopoulos
>>
>>
>>
>> On 2014-07-24, at 8:57 PM, George McNamara wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> Post images on http://www.imgur.com and include the link in your posting.
>>> *****
>>>
>>> Johannes Amon posted the URL to download a new Zeiss white paper on
>> airyscanning
>>>
>>> http://www.zeiss.com/airyscan <<a href="https://www.linkedin.com/redirect?url=http%">https://www.linkedin.com/redirect?url=http%
>> 3A%2F%2Fwww%2Ezeiss%2Ecom%2Fairyscan&urlhash=fM6m&_t=tracking_anet>
>>>
>>> Johannes post is at
>>>
>>> https://www.linkedin.com/groupAnswers?viewQuestionAndAnswers=&discussionID
>> =5898093815864004612&gid=837267
>>>
>>> and he welcomes comments.
>>>
>>>
>>> Airyscanning: A Novel Approach to Confocal Imaging
>>>
>>> Johannes Amon <<a href="https://www.linkedin.com/profile/view?id=86600201&goback=%2">https://www.linkedin.com/profile/view?id=86600201&goback=%2
>> Egde_837267_member_5898093815864004612>Online Communications and Marketing
>>>
>>> It is with great pleasure that I am able to share with you a new White
>> Paper by ZEISS. Airyscanning is a new detection concept that uses an array
>> detector to oversample each Airy disk in order to gain sensitivity,
>> resolution and speed. I'm curious to hear what you think of it! Download the
>> free White Paper here:http://www.zeiss.com/airyscan <https://www.linkedin.
>> com/redirect?url=http%3A%2F%2Fwww%2Ezeiss%2Ecom%2Fairyscan&urlhash=fM6m&_t=
>> tracking_anet>
>>>
>>>
>>> Vladimir Zhukarev and I have already commented at the linkedin group. My
>> comments are:
>>>
>>>
>>>
>>> Zeiss cites some but not all of the literature of using a (small) area
>> array to collect light that would usually be blocked by the pinhole. I'm
>> thinking that Jim Pawley had this, but could be 'off by one' (Tony Wilson,
>> Guy Cox, etc?).
>>>
>>> Too bad Zeiss and the other microscope companies have pathetically slow
>> spatial deconvolution AND on confocal microscopes deconvolution is a
>> separate post-processing step the user has to undertake manually (I managed
>> Leica confocals since 2000 and Zeiss LSM710 since 2009 ... just trained this
>> week on a Leica SP8 confocal, LAS AF 3.x, haven't seen all the features, but
>> deconvolution looks like still a separate manual step).
>>>
>>> Further, to get the maximum benefit from a confocal microscope, when
>> acquiring multiple channels, should BOTH spatially deconvolve AND spectrally
>> unmix. See the Hoppe et al 2008 Biphys J paper,
>>>
>>> http://www.ncbi.nlm.nih.gov/pubmed/18339754
>>>
>>> their 3DFSR was tested with FRET specimens but the (claimed) 10x improved
>> SNR should apply to any specimens.
>>>
>>> I also note that the Zeiss (claimed) 1.7x improvement could be massively
>> improved by acquiring 300 images (per plane) and doing 3B microscopy or SOFI
>> - for 3B see
>>>
>>> http://www.coxphysics.com/3b/
>>>
>>> for introduction to 3B (and yes, could work in 3D). 3B is also very slow.
>> Parallel processing is speeding up 3B, see
>>> http://www.optnano.com/content/2/1/7/abstract
>>>
>>> For an entry point to SOFI see
>>>
>>> http://www.optnano.com/content/2/1/2/abstract
>>>
>>> As I've noted on the Confocal Listserv
>>>
>>> http://lists.umn.edu/cgi-bin/wa?A2=ind1407&L=CONFOCALMICROSCOPY&P=5920
>>>
>>> I am psyched what parallel processing is doing (ex. NVidia TITAN Z GPU
>> card, Intel Phi 7120 card, lots of Phi's, as at https://www.tacc.utexas.edu/
>> stampede/ ) and will get even better with 2015 Knights Landing CPU's (and
>> NVidia, IBM, HP if its "The Machine" meets their goals, etc).
>>>
>>> George
>>>
>>>
>>> www.linkedin.com/in/georgemcnamara/
>>>
>>> --
>>>
>>>
>>>
>>>
>>> George McNamara, Ph.D.
>>> Single Cells Analyst
>>> L.J.N. Cooper Lab
>>> University of Texas M.D. Anderson Cancer Center
>>> Houston, TX 77054
>>> Tattletales http://works.bepress.com/gmcnamara/42"