http://www.imgur.com and include the link in your posting.
*****
Consider: How noisy is the environment? Are you in a basement or an upper
floor of a building? Is there a road with bus or truck traffic nearby? Are
there fridges or freezers or other vibration-causing equipment in the next
room?
A thicker table tends to damp out lower frequency vibrational noise, so if
you have lower frequency noise in the environment then a thicker table
would be better. A thicker and heavier table also tends to be more stable
if it is bumped or leaned on by a user, so overall thicker is better.
My own lab is on a second floors, near a major highway with truck traffic,
with a room full of -80C freezers down the hall with lots of compressors
rattling away. I went with huge active-leveling air floated 18" thick
tables and have had no issues with my confocals and two-photons re
vibration.
Craig Brideau
On Thu, Sep 11, 2014 at 5:37 AM, Deepak Nair <
[hidden email]>
wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi,
>
> We are planning to build a single molecule based super resolution system. I
> plan to put only a microscope on my active vibration isolation table. The
> weight of the optical components (including microscope)and accessories on
> the table would not exceed more than 100 Kg.
>
> Is it necessary to go for 6-8 inch thick breadboard for the active
> vibration isolation platfrom or an active vibration table with 4 inch
> breadboard be fine for the job?
>
> When I compare the graphs of the performance for the most of the active air
> based optical isolation platforms with breadboards (whether 4 inch or 8
> inch) and with anti vibration frame (cast or independent legs) perform
> well within the standard range ?
>
> Is there any specific rules for selecting active antivibration platforms ?
>
> Any input will be appreciated.
>
>
> Best
> Deepak
>
> --
> Deepak Nair
> Assistant Professor
> Centre for Neuroscience
> Indian Institute of Science
> Bangalore 560012
>