An after thought: in the confocal microscope, given a pinhole size of 0.8 - 1.0 AU, the shell thickness should appear thinest when centrally focused on.  This is in fact a good test of one’s optical system.

Dan

On Sep 19, 2014, at 2:43 PM, Knecht, David <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> We have been imaging fluorescent beads in wide field and confocal microscopes.  The beads are 3 µm in diameter and surface labeled.  In the confocal, they look like doughnuts.  My expectation was that in wide field (60-100x 1.3-1.4 NA), they would look more uniformly labeled.  In reality, they look like doughnuts.  Can someone explain why?  
>