grid target etched into borosilicate glass. They are ~20 um deep and
so I can flow different fluorophores through. The cost was very
targets. Still, they are virtually indestructable (can park the beam
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> I've been bugging various companies for years to get a good 2P and confocal
> fluoresence target developed, but thus far nobody's come up with anything
> that really works that costs less than $1000.
>
> Craig Brideau
>
>
> On Wed, Jul 23, 2014 at 11:26 AM, Michael Giacomelli <
[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> Post images on
http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Unfortunately, it is extremely difficult to record the reflection mode
>> image in our microscope, as the NIR light will not pass through the
>> dichroic, and even if we changed out the dichroic for a beam splitter,
>> out PMTs are not sensitive to NIR light.
>>
>> Translating a bead though is a very interesting idea. We use a high
>> precision thorlabs MLS stage. I will double check the repeatability,
>> but in theory I could program it to perform an extremely high
>> resolution distortion measurement. Of course, getting a single bead
>> onto a coverslip (or else programming software to automatically
>> identify the same bead across hundreds of frames) will be somewhat
>> tricky.
>>
>> Its a shame that getting targets for multiphoton is so difficult.
>> Ideal I suppose would be something etched in glass so that I could
>> flow my own fluorophore through it. I will have to look around and
>> see if I can get a lithography facility to make something like that.
>>
>> Thanks,
>> Mike
>>
>> On Wed, Jul 23, 2014 at 3:19 AM, Mark Cannell
>> <
[hidden email]> wrote:
>> > *****
>> > To join, leave or search the confocal microscopy listserv, go to:
>> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> > Post images on
http://www.imgur.com and include the link in your
>> posting.
>> > *****
>> >
>> > Thinking about this a bit more, I realised that your existing grid is
>> perfectly suitable as you don’t need 2P excitation to measure field
>> distortion at the wavelength you are using, Just turn the power down a lot
>> and record a reflection image,
>> >
>> > Cheers Mark
>> >
>> >>>
>> >>> Hi Mark,
>> >>>
>> >>> I'm assuming you mean beads deposited in a grid or something similar?
>> >>> Could you point me to where you found this?
>> >>>
>> >>> Thanks,
>> >>> Mike
>> >>>
>> >>> On Tue, Jul 22, 2014 at 3:58 AM, Mark Cannell
>> >>> <
[hidden email]> wrote:
>> >>>> *****
>> >>>> To join, leave or search the confocal microscopy listserv, go to:
>> >>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> >>>> Post images on
http://www.imgur.com and include the link in your
>> >>> posting.
>> >>>> *****
>> >>>>
>> >>>> Hi Michael
>> >>>>
>> >>>> I’ve always used beads to test the microscope. They are relatively
>> cheap
>> >>> and if you blow one up it don’t matter. As you say, evaporated metal
>> >>> targets are too easily damaged (as Jim Pauley and I discovered when
>> testing
>> >>> my microscope back in ’95!)
>> >>>>
>> >>>> HTH
>> >>>>
>> >>>> Mark
>> >>>>
>> >>>> On 22/07/2014, at 4:30 am, Michael Giacomelli <
[hidden email]> wrote:
>> >>>>
>> >>>>> *****
>> >>>>> To join, leave or search the confocal microscopy listserv, go to:
>> >>>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> >>>>> Post images on
http://www.imgur.com and include the link in your
>> >>> posting.
>> >>>>> *****
>> >>>>>
>> >>>>> Could someone recommend a good multiphoton distortion target?
>> >>>>> Preferably one that is relatively robust against damage.
>> >>>>>
>> >>>>> Currently I am using a thorlabs wire grid target with fluorophore
>> >>>>> behind it. The beam is attenuated by the metal, resulting in an
>> image
>> >>>>> of the grid. However, passing through the entire 1 mm slide greatly
>> >>>>> reduces my resolution, and unless I am extremely careful, the grid is
>> >>>>> ablated by the beam.
>> >>>>>
>> >>>>> Thanks,
>> >>>>> Michael
>> >>>>
>> >>>>
>> >>>> Mark B. Cannell Ph.D. FRSNZ
>> >>>> Professor of Cardiac Cell Biology
>> >>>> School of Physiology & Pharmacology
>> >>>> Medical Sciences Building
>> >>>> University of Bristol
>> >>>> Bristol
>> >>>> BS8 1TD UK
>> >>>>
>> >>>>
[hidden email]
>> >>>
>> >
>> >
>> >
>> > Mark B. Cannell Ph.D. FRSNZ
>> > Professor of Cardiac Cell Biology
>> > School of Physiology & Pharmacology
>> > Medical Sciences Building
>> > University of Bristol
>> > Bristol
>> > BS8 1TD UK
>> >
>> >
[hidden email]
>>