http://confocal-microscopy-list.275.s1.nabble.com/Re-Vibration-Isolation-tp7582683p7582685.html
I'm betting 10Hz was when somebody was running late... @:-)
that shoot right through the table. You can usually get frequency plots
from the manufacturers that show the response of a given table. Do keep in
>
> Anyway thanks for all your help...
>
> Deepak
>
>
> On Mon, Sep 15, 2014 at 8:26 AM, Dan Focht <
[hidden email]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Listers
> >
> > You might get a kick out of this.
> > In the 70’s I was servicing for a Zeiss dealer.
> > I was called to a hospital in Washington DC, ( of course), because their
> > new ceiling mounted surgical Eye Scope became blurry.
> > I went in and checked it out to find that when they installed their new
> > air-conditioning compressor they installed it adjacent the mounting for
> the
> > microscope!
> >
> > True story!
> >
> > Dan
> >
> >
> >
> > On Sep 14, 2014, at 8:25 PM, Smith, Benjamin E. <
[hidden email]>
> > wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > If you want to be more rigorous, ask your engineering or physics
> > department if they have a sensitive, calibrated accelerometer and a
> > spectrum analyzer. Place the accelerometer on the floor throughout the
> > room you want to place the equipment, and hook it up to the spectrum
> > analyzer. This way you will know exactly the frequencies you need to
> damp
> > out.
> >
> > Ben Smith
> > ________________________________________
> > From: Confocal Microscopy List [
[hidden email]] on
> > behalf of Deepak Nair [
[hidden email]]
> > Sent: Thursday, September 11, 2014 6:37 AM
> > To:
[hidden email]
> > Subject: Vibration Isolation
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi,
> >
> > We are planning to build a single molecule based super resolution
> system. I
> > plan to put only a microscope on my active vibration isolation table. The
> > weight of the optical components (including microscope)and accessories on
> > the table would not exceed more than 100 Kg.
> >
> > Is it necessary to go for 6-8 inch thick breadboard for the active
> > vibration isolation platfrom or an active vibration table with 4 inch
> > breadboard be fine for the job?
> >
> > When I compare the graphs of the performance for the most of the active
> air
> > based optical isolation platforms with breadboards (whether 4 inch or 8
> > inch) and with anti vibration frame (cast or independent legs) perform
> > well within the standard range ?
> >
> > Is there any specific rules for selecting active antivibration platforms
> ?
> >
> > Any input will be appreciated.
> >
> >
> > Best
> > Deepak
> >
> > --
> > Deepak Nair
> > Assistant Professor
> > Centre for Neuroscience
> > Indian Institute of Science
> > Bangalore 560012
> >
>
>
>
> --
> Deepak Nair
> Assistant Professor
> Centre for Neuroscience
> Indian Institute of Science
> Bangalore 560012
>