well. In this case it depends on what your nosepiece is threaded for. RMS
objectives like the 16X/0.8 and 25X/1.1 sold by Nikon. The back aperture of
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Maurice,
>
> Olympus objectives have a RMS thread, while Nikon uses M25. You can get
> adapters at Thorlabs (no commercial interest).
> Also the parfocal distance is different which might result in a problem
> with your z-drive.
> This is a good summary:
>
http://www.microscopyu.com/articles/optics/cfintro.html (no commercial
> interest).
> Not sure what you mean with beam expander. Maybe you mean tube lens?
> The tube lens focal length for Nikon is 200 mm while Olympus uses 180 mm.
> But that should not be a big problem.
> Both companies do their optical corrections in the objective and not the
> tube lens, I believe. So mixing Leica and Zeiss with the "Japanese"
> manufacturers might be a bigger problem. Maybe someone on the list can
> confirm?
>
> Hope that helps,
>
> Jens
>
>
> > On Nov 26, 2014, at 1:23 PM, Needham, Maurice <
[hidden email]>
> wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi Everyone,
> >
> > We have an Olympus two-photon system that we would like to use a Nikon
> 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam
> expander on the Olympus will not compensate for this lens but wondered if
> anyone has experience with “fooling” the software to allow compensation
> when using non-Olympus lenses. Also, we would appreciate comments on the
> utility of this lens for two-photon imaging as we also have a custom system
> that won’t have the beam expander problem.
> >
> > In addition, does anyone know of a supplier or accurate method of
> fabricating adaptors to allow use of objective lens between different
> manufacturer's scopes. Different objective screw thread is the main issue.
> >
> > Thanks for your help,
> >
> > Maurice
>