Hi Maurice,
Usually if both objectives are sold separately from the microscope system,
you can interchange them. If they are sold only as part of a system, they
often depend on customized tube lenses to get best performance (although
sometimes they still work well).
However, besides different thread lengths, be aware that different
objectives, even with the same magnification, may have very different back
focal plane locations. Before you swap the objectives, request the back
focal plane location for the old objective and the new from the
manufacturers and figure out if you can mechanically make it work in your
system (that is, the new objective won't have its back focal plane such
that its threading would be in the middle of your dichroic).
Assuming you can make it work, thorlabs is great for adapters.
Mike
On Wed, Nov 26, 2014 at 1:23 PM, Needham, Maurice <
[hidden email]>
wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Everyone,
>
> We have an Olympus two-photon system that we would like to use a Nikon 16X
> lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam
> expander on the Olympus will not compensate for this lens but wondered if
> anyone has experience with “fooling” the software to allow compensation
> when using non-Olympus lenses. Also, we would appreciate comments on the
> utility of this lens for two-photon imaging as we also have a custom system
> that won’t have the beam expander problem.
>
> In addition, does anyone know of a supplier or accurate method of
> fabricating adaptors to allow use of objective lens between different
> manufacturer's scopes. Different objective screw thread is the main issue.
>
> Thanks for your help,
>
> Maurice
>