Re: inter-manufacturer objective compatibility

Posted by Needham, Maurice on
URL: http://confocal-microscopy-list.275.s1.nabble.com/inter-manufacturer-objective-compatibility-tp7582994p7583007.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Everyone,
                    Thank you so much for all your responses. I have some homework to do but i'm sure I will get it sorted due to all your helpful comments. This is my first post to the Microscopy list and i'm very impressed with the response. Its comforting to have access to such a wealth of knowledge on Imaging. Take care,

with thanks,

Maurice
________________________________________
From: Confocal Microscopy List [[hidden email]] on behalf of Steffen Dietzel [[hidden email]]
Sent: Thursday, November 27, 2014 5:54 AM
To: [hidden email]
Subject: Re: inter-manufacturer objective compatibility

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Maurice,

we have a LaVision Biotec multi-photon which is built around an Olympus
stand. A few years ago bought additional objectives, to supplement the
(very good) 20x 0.95 Objective from Olympus for specific tasks. After
testing several ones we ended up buying one from Leica (40x, 1.1, better
resolution, but flat top) and the 16x Nikon (steeper angle, could access
places the others couldn't). Putting all theory aside, they both work
nicely for their specific purpose.

Steffen

Am 26.11.2014 19:23, schrieb Needham, Maurice:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Everyone,
>
> We have an Olympus two-photon system that we would like to use a Nikon 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with.  We realize that the beam expander on the Olympus will not compensate for this lens but wondered if anyone has experience with “fooling” the software to allow compensation when using non-Olympus lenses.  Also, we would appreciate comments on the utility of this lens for two-photon imaging as we also have a custom system that won’t have the beam expander problem.
>
> In addition, does anyone know of a supplier or accurate method of fabricating adaptors to allow use of objective lens between different manufacturer's scopes. Different objective screw thread is the main issue.
>
> Thanks for your help,
>
> Maurice
>


--
------------------------------------------------------------
Steffen Dietzel, PD Dr. rer. nat
Ludwig-Maximilians-Universität München
Walter-Brendel-Zentrum für experimentelle Medizin (WBex)
Head of light microscopy

Mail room:
Marchioninistr. 15, D-81377 München

Building location:
Marchioninistr. 27,  München-Großhadern