Hi
The only problem I foresee will be the varying angle to the dichroic which will change its cutoff over the scan area. The effect may be small and you may get away with it.
HTH
Mark
On 29/11/2014, at 4:23 pm, Heping Yuan <
[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Hi everyone, I was looking to modify an old confocal system into a two-photon system. I
> had two questions and would appreciate any help.
>
> 1) A section of the schematic of typical two-photon is as follows
> (
http://www.thorlabs.com/tutorials.cfm?tabID=32729):>
> Galvos -> Scan Lens -> Tube Lens -> Dichroic (690 LP) -> Objective
>
> where fluorescence emission is reflected off Dichroic towards a collection lens and PMT's
>
> I'm wondering if there are some unforeseen problems in changing the order as follows:
>
> Galvos -> Scan Lens -> Dichroic (690 LP) -> Tube Lens -> Objective
>
> where the tube lens can focus the fluorescence emission back to the Dichroic and directly
> into the PMT's (without a collection lens).
>
> 2) What is a common procedure to image the back aperture of the objective to the input
> window of the PMT? My first thought is to create a collimated source with size > than
> back aperture and shine directly into the back aperture with objective off (to see spot size
> at PMT input). Would this be correct? If so, practically speaking what type of source is
> typically used?
>
> Thanks,
> Heping
Mark B. Cannell Ph.D. FRSNZ
Professor of Cardiac Cell Biology
School of Physiology & Pharmacology
Medical Sciences Building
University of Bristol
Bristol
BS8 1TD UK
[hidden email]