particularly for larger fields of view. If you take apart the thorlabs
back of the objective. This maximizes light collection.
of your objective. I would measure the displacement you intend to add
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Short answer to (1) is, astigmatism.
>
> Transmission through a flat, tilted piece of glass gives only a lateral
> deflection to a collimated beam, but it gives astigmatism to a focusing
> beam. You could imagine using a dichroic that reflects the excitation
> rather than transmitting it, but be careful about flatness (most dichroics
> are quite curved, sadly) which can also give astigmatism.
>
> Because of some bad decisions, I ended up using a dichroic in the same
> manner you describe. You can cancel a lot of the resulting astigmatism with
> a second piece of glass tilted the same amount but rotated 90 degrees about
> the optic axis. On the other hand, why bother solving a problem you don't
> have to have?
>
> On Sat, Nov 29, 2014 at 11:23 AM, Heping Yuan <
[hidden email]>
> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi everyone, I was looking to modify an old confocal system into a
> > two-photon system. I
> > had two questions and would appreciate any help.
> >
> > 1) A section of the schematic of typical two-photon is as follows
> > (
http://www.thorlabs.com/tutorials.cfm?tabID=32729):> >
> > Galvos -> Scan Lens -> Tube Lens -> Dichroic (690 LP) -> Objective
> >
> > where fluorescence emission is reflected off Dichroic towards a
> collection
> > lens and PMT's
> >
> > I'm wondering if there are some unforeseen problems in changing the order
> > as follows:
> >
> > Galvos -> Scan Lens -> Dichroic (690 LP) -> Tube Lens -> Objective
> >
> > where the tube lens can focus the fluorescence emission back to the
> > Dichroic and directly
> > into the PMT's (without a collection lens).
> >
> > 2) What is a common procedure to image the back aperture of the objective
> > to the input
> > window of the PMT? My first thought is to create a collimated source with
> > size > than
> > back aperture and shine directly into the back aperture with objective
> off
> > (to see spot size
> > at PMT input). Would this be correct? If so, practically speaking what
> > type of source is
> > typically used?
> >
> > Thanks,
> > Heping
> >
>