http://confocal-microscopy-list.275.s1.nabble.com/Two-photon-microscope-questions-tp7583010p7583020.html
sides is inside the objective). There should be negligible signal hitting
microscope if any part is not fully light tight. I think this is the main
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Dear Heping,
>
> To your 2nd question:
> >> 2) What is a common procedure to image the back aperture of the
> objective to the input window of the PMT?
> >> My first thought is to create a collimated source with size > than
> >> back aperture and shine directly into the back aperture with objective
> off (to see spot size
> >> at PMT input). Would this be correct?
>
> This sounds difficult to me. The detector size is usually so big that it
> would be difficult to find the optimal alignment by this method.
>
> For a quick alignment, I would, if the excitation light path is perfectly
> aligned, put a homogeneous plastic fluoslide under the objective and align
> the detection pathway such that the picture you get looks as good as
> possible. Do this with the lowest zoom setting (highest scanning angle),
> otherwise it will always look perfect. With "good", I mean a FOV with the
> highest brightness in its center. Simply play around with mirror angles
> while imaging and check the picture live. Usually, due to aperture clipping
> effects while scanning, the FOV is not totally homogeneous. This also
> depends on the highest zoom factor allowed by your scanners or your
> software. -- If someone has a better idea, I'd be happy to learn them!
>
> As already mentioned, I would also vote for putting the dichroic as close
> to the objective as possible.
>
>
> As a sidenote, I have always been wondering why some 2P-microscopes I've
> seen use such long detection pathways (>15 cm) for 2P microscopy, with all
> the tubes and inner parts being completely black. Wouldn't it be much
> better to coat the full detection pathway surfaces with an aluminium or
> silver coating in order to trap every scattered photon until it impinges on
> the detector? Has anyone tried this out and compared it to non-coated
> surfaces for highly scattering conditions?
> Best,
> Peter
>