Yes,
the BFP of the objective should be the only aperture that limits your
optical system. By properly imaging this onto the detector you don't loose
any useful light. It is convenient to use black tubes, as any light leaking
into the system will degrade your image considerably.
With a high magnification objective (say 60x 1.4NA) the rays between the
objective and the tube lens are usually well collimated, so you don't need
to be concerned with 'long' relay lenses between the objective and the non-
descanned detector (this is not entirely true for scattered two-photon
fluorescence, you need to perform some sort of simulation or calculation to
get an idea about the diameter/focal length of your relay lenses necessary
to collect out-of-focus light) .
Best, Zdenek, www.kcci.virginia.edu
---------- Původní zpráva ----------
Od: Michael Giacomelli <
[hidden email]>
Komu:
[hidden email]
Datum: 29. 11. 2014 18:27:33
Předmět: Re: Two-photon microscope questions
"*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Hi Peter,
With good optical design you can collect nearly the full radiance of the
objective back aperture (that is, the only place where light will hit the
sides is inside the objective). There should be negligible signal hitting
the lens tubes, but perhaps a bit of light leakage from outside the
microscope if any part is not fully light tight. I think this is the main
concern, although in practice I'm not sure how much it really matters.
Mike
On Sat, Nov 29, 2014 at 6:07 PM, Peter Rupprecht <
[hidden email]>
wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Dear Heping,
>
> To your 2nd question:
> >> 2) What is a common procedure to image the back aperture of the
> objective to the input window of the PMT?
> >> My first thought is to create a collimated source with size > than
> >> back aperture and shine directly into the back aperture with objective
> off (to see spot size
> >> at PMT input). Would this be correct?
>
> This sounds difficult to me. The detector size is usually so big that it
> would be difficult to find the optimal alignment by this method.
>
> For a quick alignment, I would, if the excitation light path is perfectly
> aligned, put a homogeneous plastic fluoslide under the objective and align
> the detection pathway such that the picture you get looks as good as
> possible. Do this with the lowest zoom setting (highest scanning angle),
> otherwise it will always look perfect. With "good", I mean a FOV with the
> highest brightness in its center. Simply play around with mirror angles
> while imaging and check the picture live. Usually, due to aperture
clipping
> effects while scanning, the FOV is not totally homogeneous. This also
> depends on the highest zoom factor allowed by your scanners or your
> software. -- If someone has a better idea, I'd be happy to learn them!
>
> As already mentioned, I would also vote for putting the dichroic as close
> to the objective as possible.
>
>
> As a sidenote, I have always been wondering why some 2P-microscopes I've
> seen use such long detection pathways (>15 cm) for 2P microscopy, with all
> the tubes and inner parts being completely black. Wouldn't it be much
> better to coat the full detection pathway surfaces with an aluminium or
> silver coating in order to trap every scattered photon until it impinges
on
> the detector? Has anyone tried this out and compared it to non-coated
> surfaces for highly scattering conditions?
> Best,
> Peter
>"