Dear Anton,

Just a quick idea: the plastic fluorescent slides distributed by Chroma
(but also others) can be also bleached (if you have a somewhat stronger
laser). Could this work for you?

Greetings    Gabor

On 12/1/14 2:40 PM, "Anton Kamnev" <[hidden email]> wrote:

>*****
>To join, leave or search the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>Post images on http://www.imgur.com and include the link in your posting.
>*****
>
>Dear all,
>
>In our LM core we face a problem of frequent calibration of our "FRAP"
>devices. Although the calibration itself is quite fast, many users just
>don't have good samples which could be used for this purpose.
>
>Therefore, I was wondering if there are any commercial products for
>calibration of bleaching devices? What I'm looking for is a slide with a
>thin uniform layer of a dye which could be bleached. The dye should be
>immobilized so a bleaching pattern would be observable for some period of
>time after the bleaching.
>
>I would really appreciate if you could share some experience about samples
>you use for calibration of bleaching patterns.
>
>Although I'm primarily interested in commercial solutions, I would very
>much
>appreciate some protocols for home-made calibration slides as well.
>
>Thanks in advance!
>Anton.
>
>---------------------------
>Anton Kamnev, PhD
>Imaging Manager
>Mechanochemical Cell Biology Building
>Division of Biomedical Cell Biology
>Warwick Medical School
>The University of Warwick
>Coventry, CV4 7AL UK
>
>tel: +44 (0) 24-7615-1934
>cell: +44 (0) 782-408-6941
>email: [hidden email]
>
>Divisional Secretary: Karl Waldron
>Tel: +44(0) 24-7615-1154
>[hidden email]