might want to check out this article. A PDF is not very good quality
Questions"? If not, please give us your input. Every data point is
important! Deadline: Dec 6. "Thank you gift": A chance to win one of
>*****
>To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>Post images on
http://www.imgur.com and include the link in your posting.
>*****
>
>Non-commercial response:
>
>Jeff,
>
>In my experience, use a Planapochromat 60/1.4 or 100/1.45 oil
>objective, 1.4NA oil condenser, precision interference line green
>filter(546/10), and "standard" or "high" contrast DIC objectives &
>DIC condenser prisms. You will need to carefully test which
>combination of DIC prisms work best for you. Be sure the back
>aperture of the objective is completely filled. You will probably
>need to introduce a 1.5x-2.5x magnification factor in the camera
>light path, and, in this case, a solid state illuminator should give
>you more light than a 100W halogen lamp. Some microscope systems
>offer a high NA illuminator adapter that will accept a solid state
>illuminator liquid light guide, and this works well for this
>application. You will probably need to adjust your image histogram
>for the best image.
>
>I recently used the above combination on an inverted microscope and
>achieved excellent results with a 5MP sCMOS camera.
>
>David
>
>David J. Claypool
>Digital Imaging Product Manager
>Micro Video Instruments
>Office: 800-875-2041 x5221
>Cell: 603-809-5342
>
[hidden email]
>
> > On Nov 28, 2014, at 9:30 PM, "Jeff Spector" <
[hidden email]> wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your posting.
> > *****
> >
> > Greetings,
> > This isn't exactly a "confocal" questions but I know a lot of "micoscopy
> > gurus" live on this list so I thought it a good place to ask this. I have a
> > colleague who is trying to image individual (i.e. small and diffraction
> > limited) microtubules in a flow chamber by using DIC. They are currently
> > using a 100x 1.45 oil Objective, but only a .52 LWD condenser. They were
> > using a solid state light source but couldn't get good image so we switched
> > to a lamp for illumination and the images are much better, and we can now
> > see the microtbules but there still isn't a lot of contrast. My question
> > is, is it worth it to go to a high NA (perhaps oil immersion) condenser,
> > and can anyone think of why the lamp would give a better DIC image than a
> > solid state light source?
> > thanks in advance for the help...
> > -Jeff