independent of the fixation itself, you should also have an eye on your
embedding medium. Hardening medium such as Mowiol or ProLong tends to
flatten the cells quite a bit, compared to non-hardening such as
Vectashield.

Steffen

Am 09.12.2014 11:52, schrieb Jeremy Adler:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Cells on coverslips tend to flatten during fixation.
> In LM the Z resolution is always poorer than in XY and shrinkage in Z exacerbates this mismatch.
> Does anyone have any suggestions for preventing or reducing shrinkage.
>


--
------------------------------------------------------------