crosstalk is quite nice. Devils advocate arguments against going fast with
1. 2p cross sections are very very low compared to 1p; it takes a lot of
levels pretty fast (>1 W average power). Even though IR light is absorbed
mean non-negligible heating. Getting the same degree of parallelization as
2. Typical pulse rates for 2p (>10 ns) are long compared to fluorescent
3. I'm pretty sure you get fewer signal photons per bleaching event with 2p
compared to 1p, when imaging a single plane. Can anyone confirm/deny? I
4. I'm not even sure you can saturate excitation with 2p, compared to 1p.
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Multi-beam multiphoton (eg LaVision Biotec) also limits bleaching to the
> focal plane and has the advantage over spinning disk confocal that there is
> no cross-talk. No commercial association, but I do know a very satisfied
> user.
>
> Guy
>
> Guy Cox, Honorary Associate Professor
> School of Medical Sciences
>
> Australian Centre for Microscopy and Microanalysis,
> Madsen, F09, University of Sydney, NSW 2006
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:
[hidden email]]
> On Behalf Of James Pawley
> Sent: Sunday, 4 January 2015 11:47 AM
> To:
[hidden email]
> Subject: Re: High speed spinning disc confocal with EMCCD camera -
> commercial response
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> >*****
> >To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> >Post images on
http://www.imgur.com and include the link in your posting.
> >*****
>
>
>
> Details aside, data rate will always be proportional to how much light is
> detected/second. More beams will produce more data/second.
> Single beam instruments really can't compete because they intensity in a
> focused confocal spot is already close to singlet-state saturation. But the
> quality of the data will vary between techniques.
> What do you "need to see"?.
>
> I would bet on light sheet/SPIM. Damage only in the illuminated plane and
> simple optics to a (effective) high-QE EM-CCD or sCMOS camera.
>
> JP
>
> >Hi all,
> >
> >Does anyone think it would be possible to tabulate a 'speed limit' for
> >the various options discussed? I know it sounds near impossible to
> >come up with a standard basis for comparison, but let's say something
> >approximating a 512x512 acquisition either fixed or or a volume that
> >includes 10 z steps (e.g., using a piezo stage when relevant). It
> >would be great to have an order of magnitude idea how to compare
> >technologies like a resonant scanner, Optera-type swept field scanner,
> >spinning disc, VCS super-spinning disc or light sheet instrument when
> >FPS is a major priority and excitation light is not limiting. Maybe we
> >could crowdsource it from what users actually get in practice.
> >
> >All the best,
> >
> >
> >Tim
> >
> >Timothy Feinstein, Ph.D. | Manager, Core for Confocal Microscopy and
> >Quantitative Imaging
> >333 Bostwick Ave., N.E., Grand Rapids, Michigan 49503
> >Phone: 616-234-5819 | Email:
[hidden email]
> >
> >
> >
> >
> >
> >
> >
> >On 12/30/14, 2:36 AM, "Andrea Latini" <
[hidden email]> wrote:
> >
> >>*****
> >>To join, leave or search the confocal microscopy listserv, go to:
> >>
http://scanmail.trustwave.com/?c=129&d=09ai1Hwf389A_JSBGnaEcBqKN1nFKNq> >>OL1R
> >>ax-qpJw&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfoca
> >>lmic
> >>roscopy
> >>Post images on
> >>
http://scanmail.trustwave.com/?c=129&d=09ai1Hwf389A_JSBGnaEcBqKN1nFKNq> >>OLwF bleD9dw&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link in
> >>your posting.
> >>*****
> >>
> >>Dear Andrew,
> >>the VCS (Video Confocal Super Resolution), module is an X-Light
> >>Spinning disk system add-on.
> >>the disk is out of the optical path when in VCS mode (i.e. 'bypass'
> mode).
> >>basically, it's a new implementation of structured illumination
> >>technology aimed to fast image acquisition with no resolution
> >>limitations that are spinning disk related.
> >>
> >>I'll be pleased to discuss more, please get in touch.
> >>
> >>Regards.
> >>
> >>Andrea
> >>
[hidden email]
> >>
> >>
> >>On Mon, 29 Dec 2014 16:58:15 -0500, Andrew York
> >><
[hidden email]> wrote:
> >>
> >>>*****
> >>>To join, leave or search the confocal microscopy listserv, go to:
> >>>
http://scanmail.trustwave.com/?c=129&d=09ai1Hwf389A_JSBGnaEcBqKN1nFKN> >>>qOL1
> >>>Rax-qpJw&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfo
> >>>calm
> >>>icroscopy
> >>>Post images on
> >>>
http://scanmail.trustwave.com/?c=129&d=09ai1Hwf389A_JSBGnaEcBqKN1nFKN> >>>qOLw FbleD9dw&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link
> >>>in your posting.
> >>>*****
> >>>
> >>> Is there information available about this product? Is this an
> >>>implementation of Enderlein's spinning disk paper? Also, 80 nm seems...
> >>>optimistic? Is this with very short wavelength light, or just a
> >>>slightly different definition of resolution than I'm used to?
> >>>
> >>>On Mon, Dec 29, 2014 at 4:10 PM, Andrea Latini <
[hidden email]>
> >>>wrote:
> >>>
> >>>> *****
> >>>> To join, leave or search the confocal microscopy listserv, go to:
> >>>>
> >>>>
http://scanmail.trustwave.com/?c=129&d=09ai1Hwf389A_JSBGnaEcBqKN1nFK> >>>>NqOL
> >>>>1Rax-qpJw&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dcon
> >>>>foca
> >>>>lmicroscopy
> >>>> Post images on
> >>>>
http://scanmail.trustwave.com/?c=129&d=09ai1Hwf389A_JSBGnaEcBqKN1nFK> >>>>NqOL wFbleD9dw&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link
> >>>>in your
> > >>>posting.
> >>>> *****
> >>>>
> >>>> - commercial response
> >>>>
> >>>> thanks for reporting your experience with our Confocals Marco.
> >>>>
> >>>> the new Video Super Resolution module for XLight allows for 50ms
> >>>>exposure
> >>>> time and <1
> >>>> second, 80nm spatial resolution; this is possible with large Cuda
> >>>> programming we've been
> >>>> developing during past months and introduced @SfN 2014 as a product.
> >>>>
> >>>> soon on our website and in your Lab, hopefully!
> >>>>
> >>>> Cheers.
> >>>>
> >>>> Andrea
> >>>>
> >>>> CrestOptics
> >>>>