>@Zdenek: A SNR of 3.5 or a little bit more might be enough for some
>applications ... I was planning to measure the photon count per
>pixel (using this method :
>
http://labrigger.com/blog/2010/07/30/measuring-the-gain-of-your-imaging-system/
>), but I always was busy with other things, so I cannot give you
>numbers for my imaging system.
Hi all,
Thanks to Labrigger for working on this important topic.
However, I have read his analysis and think that the assumption that
one can use this procedure to measure the number of photoelectrons
(PE: i.e., detected photons) created at the photocathode (PC) of the
PMT may be an over-simplification.
The analysis depends on the assumption that the only source of noise
in the data recorded in the "image" of a flat white field is Poisson
Noise associated with the small number of PEs produced at the
photocathode. This might be true if PMTs were free from
multiplicative noise but in fact Poisson Noise also affects every
stage in the multiplication of a single PE after it leaves the PC. In
the very unusual case that the voltage between the PC and the first
dynode is 500-600 volts (and that this dynode has both the optimal
shape and the best GaAs surface), the gain of this stage may be 25
+/-5 or 20% additional noise. More commonly, this gain will be closer
to 4 +/-2 or 50% additional noise. More noise is added at each stage
and even though these noise terms are not additive (they are combined
as the sqrt of the sum of the squares), it is not at all uncommon for
this process to double or even triple the variation present in the
resulting signal beyond what one would expect from Poisson Noise
applied only to the number of PE. Furthermore, this added noise will
be somewhat larger if the system is working at a relatively high
signal level because then the PMT will be turned down, the gain/stage
correspondingly lower and the Poisson Noise proportionally higher.
Offsetting this error to some extent is the finite bandwidth of the
entire amplifier system (PMT plus the electronics between the final
dynode and the ADC). This bandwidth is in general unknown but may be
adjusted by the computer to more-or-less match what the computer
estimates is needed to pass the finest optical details that the
system can transmit on the basis of settings for wavelength,
objective NA, zoom/pixel size, and even PMT setting (high PMT voltage
implies a noisy signal that may benefit from the artificial,
1-dimensional smoothing that attends lower bandwidth).
Clearly, because bandwidth limits the maximum excursion that can be
transmitted between one pixel and its neighbour, it will tend to
reduce the apparent noise present in the digitized signal. The
magnitude of this clipping is unknown but may vary with the
parameters mentioned above.
This is relevant because, unlike the optical signal, the Poisson
Noise signal that we are searching for shows no correlation between
adjacent pixels. In particular, following the blog's suggestion of
using a high zoom (to reduce fixed pattern noise) may cause the
computer to limit the bandwidth more than using a lower zoom.
Although, as noted above, because these two factors bias the results
in opposite directions, their effects may cancel each other out to
some extent. However, we need to know a lot more about how the
components are actually operating before we can decide whether and to
what extent this is true.
The analysis also assumes that there is no fixed patterns noise in
the image of a "flat white field" as might be caused, for instance,
by field curvature, spherical aberration, vignetting, dust or other
optical parameters that may change detected signal across the field
of view. I note that many of these sources of non-Poisson Noise can
be substantially reduced by recording two sequential frames and
obtaining a measure of the noise by subtracting one from the other.
For the analysis to work, it is also important to set the brightness
control (DC - offset) so that zero signal corresponds to closely to
zero intensity in the image memory.
I should note that multiplicative noise ceases to be a factor in
systems employing either hybrid PMT (where the first stage gain is
about 10,000) or effective photon-counting (i.e. a photon counting
where the recorded peak pixel signal is at least 10x smaller than the
saturation count rate of the system as set by pulse-pileup.).
One can avoid multiplcative noise by recording the data using a CCD
(but NOT on an EM-CCD used with the electronic gain turned on) and
the record-two-then-subtract approach can again be used to reduce
inevitable fixed pattern noise. However, this sensor will probably
work best when recording a fairly large signal (at least 10% of
peak?) so that read noise will be relatively insignificant. And as
above, the results will again be limited by the finite bandwidth of
the FET amplifier between the read-node and the ADC. Finally, when
using a CCD for quantitative measurements, it is particularly
important to remember that they are usually set up so that zero light
corresponds to 20-50 computer intensity units.
The noise performance of sCMOS detectors is both non-Gaussian and
depends strongly on the extent to which the internal pixel-by-pixel
variations in gain and offset are detected and corrected. This will
make their use for this type of measurement somewhat more difficult
unless the signal levels are well away from the noise floor.
Bottom line: Although the procedure may indeed give a useful
benchmark that we might call the "effective gain" of the signal path,
the measurement is subject to influence by a number of imaging
parameters and will not really allow one to measure how many
recorded-signal-intensity-units correspond to one PE.
Jim Pawley
--
****************************************