Dear Carlos,
I can only give you feedback on the HyDs as we are using them on our SP5s.
You have to be more careful as on normal PMTs as the HyDs should have as few overloads as possible. Even though they have a protection against it, they are quite vulnerable.
Other than that they should be also quite reliable.
However, so far all of our 4 HyDs had to be exchanged at least once a year (in warranty) due to functional problems ( "comets" on the image), which should be solved on the new HyDs delivered more recently. So far we do not have any experience with GAsP-detectors...
I hope that helps,
Oliver
Sent from a mobile device.
Oliver Biehlmaier, PhD | Head of Imaging Core Facility | Biozentrum, University of Basel | Klingelbergstrasse 50/70 | CH-4056 Basel
Phone: +41 61 267 20 73<tel:+41%2061%20267%2020%2073> | Email:
[hidden email]<mailto:
[hidden email]> |www.biozentrum.unibas.ch<
http://www.biozentrum.unibas.ch/> |www.microscopynetwork.unibas.ch<
http://www.microscopynetwork.unibas.ch/>
Am 30.03.2015 um 08:30 schrieb Carlos Sanchez Martin <
[hidden email]<mailto:
[hidden email]>>:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Hi, good morning.
Just a quick question, please, although it may be difficult to answer yet.
I've been said that GaAsP and/or HyD detectors might have shorter lifetimes
than conventional PMTs, so that they might even be considered fungible
materials. I am assuming a proper use of that detectors.
Is this true or just a rumour?
Thank you very much in advance for your help.
Regards,
Carlos Sánchez Martín
Servicio de Microscopía Óptica y Confocal (SMOC) (Lab. 310)
(Optical and Confocal Microscopy Facility)
Centro de Biología Molecular Severo Ochoa (UAM-CSIC)
C/Nicolás Cabrera, 1. Universidad Autónoma de Madrid
Cantoblanco. E-28049. Madrid. Spain
Tlf. 34-91 196 4613/4643
Fax. 34-91 196 4420
E-mail: <mailto:
[hidden email]/
[hidden email]>
[hidden email]<mailto:
[hidden email]>/
[hidden email]
Blog: <
http://csanchezmad.wordpress.com/>
http://csanchezmad.wordpress.comWeb: <
http://www.cbm.csic.es/confocal>
http://www.cbm.csic.es/confocalRed Española de Microscopía Óptica Avanzada (REMOA)
(Spanish Network of Advanced Optical Microscopy)
<
http://remoa.wikispaces.com/>
http://remoa.wikispaces.com<
http://remoa.net/>
http://remoa.netPlataforma de Microscopía para Biociencias
Red de laboratorios de la Comunidad de Madrid
<
http://www.madrimasd.org/Laboratorios/plataformas-red/ficha.asp?IdPreli=3>
http://www.madrimasd.org/Laboratorios/plataformas-red/ficha.asp?IdPreli=3