Whats the slowest translation time (per frame) you'd be happy with?
100 ms is not very hard for a several mm deflection and a good DC
brushless microscope stage.  Faster than that over large distances can
be more difficult.

Keep in mind that settling time for samples in liquid can be
non-negligible, and there is often there is some latency associated
with triggering a microscope (especially for confocal systems).  There
is little point in getting a stage with a 50 millisecond step response
if your media keeps flowing for 100 ms after each move.  This can be a
very big factor in system throughput, even compared to actuator
response time.

Mike

On Tue, Apr 21, 2015 at 12:02 PM, Csúcs  Gábor
<[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear All,
>
> For a special application we are searching for the "fastest possible"
> motorized X-Y stage for an inverted microscope. As we would like so scan
> well-plate sized samples, the scanning range should be quite large
> (approx. 90x130 mm). On one of our dedicated high-throughput screening
> microscopes we have a special (very fast) voice-coil driven system - I was
> wondering if anyone of you know about the availability of this solution
> for normal inverted microscopes. Or do you know alternatives? I'd
> appreciate any feedback.
>
> Greetings     Gabor
>
> Gabor Csucs
> ScopeM
> Wolgang-Pauli Str. 14
> CH-8093 Zurich
>
> Phone: +41 44 633 6221