Re: PSF asymetry

Posted by Weis, Michael-2 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/PSF-asymetry-tp7583681p7583691.html

I recently had this same problem with one objective on a new confocal installation here. After acquiring a XYZ stack documenting the lateral shift of the PSF I rotated the objective 90 degrees and acquired another stack. The lateral shift rotates the same as the objective rotates therefore the defect is in the objective. The manufacturer is replacing the objective.

Cheers, Michael


Michael Weis
Supervisor,                                                  Superviser,
Microscopy Facility                                   Installation de microscopie
Science and Technology Branch          Direction générale des sciences et de la technologie 
Pacific Agri-Food Research Centre     Centre de recherches agroalimentaire du Pacifique
P.O. Box 5000, 4200 Highway 97          Boîte postale 5000, 4200 Autoroute 97
Summerland, BC, V0H 1Z0, Canada    Summerland, CB, V0H 1Z0, Canada



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier
Sent: April-21-15 1:10 AM
To: [hidden email]
Subject: PSF asymetry

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear microscopists,

I am using a very nice TIRF microscope with a 100X, 1.49 NA objective. I had the impression that there was a slight lateral shift when defocusing up and down, so I checked the PSF with 100 nm beads on a HR #1.5 coverslip.
What appears on the attached image (three planes taken at -1, 0 and +1 um) is that the PSF is not rotationnally symetric, i.e.more intense on the left-bottom side :

https://drive.google.com/file/d/0B_JeGjE7nBHWWFViM0pwSy0xR3M/view

This asymetry is quite constant over the field of view  (it is not radial relative to the center of the field). It does not depend on the illumination (it is the same under azimutal laser, TIRF laser, epifluorescence lamp). It does not depend on the filter cube used. Finally (and this is what surprises me the most), I got another brand new 100X,
1.49 objective for testing and it still shows up (the attached image is taken with the new objective).

Do you have an idea if what could be wrong, and how to correct it? Could it be caused by an internal lens? By the sample used?

Thanks for your help,

Christophe