Dear Christophe,
When I started here I was testing PSFs on one of the scopes and was getting dramatic asymmetry in several objectives. After much hunting and testing it turned out that the objective turret was cross-threaded. You might just try unscrewing your whole objective turret and reseating it.
Best of luck,
Kira