Re: Preamplifier for fast point-scanning
Posted by
Peter Rupprecht on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Re-Preamplifier-for-fast-point-scanning-tp7583751p7583758.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Thanks for the replies to Michael, Craig, an off-list - if I'm again in the situation of choosing a preamp, I will certainly have a look again at the Thorlabs model. The input noise level actually seems to be slightly lower than for the DHCPA-100 (used at "high gain", 80 MHz bandwidth). The main advantage of the Femto preamp seems to be that gain and bandwidth can be varied. Also, I sometimes used the possibility to switch to AC-coupling, which is not available for the Thorlabs model.
Peter