Re: chromatic aberration VIS vs near-IR and the Non-Descanned Detectors

Posted by Michael Giacomelli on
URL: http://confocal-microscopy-list.275.s1.nabble.com/7th-Course-on-Optical-Microscopy-Imaging-for-Biosciences-20-24-April-2015-IBMC-Porto-tp7583912p7583945.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

As Craig says, for two photon imaging with non-descanned detectors, only
the chromatic aberration of the excitation (in the NIR) matters.  This can
be quite severe as most objectives (even NIR enhanced ones) are
not achromatized in the NIR, and so can have very large chromatic
aberration for multiphoton applications.  Zeiss, for instance, recommended
us an NIR enhanced 0.45 NA objective that measured a gigantic 100 nm of
chromatic focal shift per 1 nm shift of center wavelength from 700-900 nm.
Such an objective will work fine with typical 150 fs excitation, but is
almost useless for shorter pulses or confocal imaging in the NIR.

If you think you need good broadband NIR-vis achromatization (and most
people don't really) then I would strongly recommend you request an
evaluation objective from the vendor to test before you buy it.  Many
vendors will happily recommend NIR objectives that are actually quite poor
in the NIR.

Mike

On Mon, Jun 29, 2015 at 6:44 PM, Craig Brideau <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Ah, you are using far red dyes with a confocal. That's a relatively new
> thing. In this case the NIR corrected lenses normally used for 2-photon are
> actually useful in a confocal situation since you are working with very red
> light both for excitation and emission. If you also want to image GFP then
> your system has to be very broadly corrected.
> Also, hopefully someone can chime in here, how sensitive are the Perkin
> Elmer APDs to 800 to 900nm light? Many detectors are very insensitive this
> far down. You typically want to use a multi-alkali PMT for those sorts of
> ranges. I find the Hamamatsu 10721-20 works well out to about 850-900nm yet
> is still sensitive 'enough' in the blue and green as well (I use it for GFP
> or DAPI and very red things). Here's the product sheet:
>
> http://www.hamamatsu.com/eu/en/product/alpha/P/3003/3044/H10721-20/index.html
> Getting back to your objective lens, in this case you need something that
> is very broadly chromatically corrected. You will want to consult your
> vendor as to which lenses they offer that meet this need. You will also
> want to avoid working on the edge of your field of view, and might have to
> crop down a bit as chromatic aberration gets worse near the edge of your
> full field.
>
> Best of luck!
>
> Craig Brideau
>
>
> On Mon, Jun 29, 2015 at 4:27 PM, Vitaly Boyko <[hidden email]>
> wrote:
>
> > Hi Craig,
> >
> > thank you for your reply. How would one image Alexa 790/Cy7 together with
> > let say CeFP/dapi, GFP through the one photon confocal system (with
> Perkin
> > Elmer APDs as detectors). Are there many IR APOCHROMAT lens that are
> > corrected between 400 nm (tough, I guess, at least 500 nm) and 900nm at
> 40x
> > NA 1.0-1.1 or 60x NA 1.2?Multi-photon collect scattered (non-linear,
> > non-chroma corrected) light via NDDs...
> >
> > Vitaly
> >
> >
> >
> >
> >   On Monday, June 29, 2015 6:16 PM, Craig Brideau <
> [hidden email]>
> > wrote:
> >
> >
> > Your question is fairly vague, but here are some general answers: With 2
> > photon the important thing is how well the laser is focused. This means
> > your objective has to be well corrected for the NIR beam from the laser.
> > The resulting visible fluorescence is directed towards wide-aperture
> PMTs,
> > which don't care about aberration very much.
> > For single-photon confocal, you have to get your fluorescence through the
> > confocal pinhole, which will be *very* picky about aberration. Since you
> > are trying to get the light through a very small aperture any aberration
> > will cause signal loss.
> >
> > Craig
> >
> > On Mon, Jun 29, 2015 at 4:04 PM, Vitaly Boyko <[hidden email]>
> > wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Dear List,
> > I would greatly appreciate if someone could shed light on the extent of
> > lateral and axial chromatic aberrations correction, and its limitations
> in
> > the VIS portion of spectrum when image has been projected into the
> > non-descanned detectors (one and two photon imaging)?
> > Many thanks in advance.
> > Vitaly
> >
> >
> >
> >
> >
> >
> >
> >
>