Unfortunately, the link didn't work.

Shouldn't it be 1.22?

Normally, you see it written as  0.6 only if it has been divided by 2NA.

Barbara Foster,  President and  Chief Strategic Consultant

The Microscopy & Imaging Place, Inc.
7101 Royal Glen Trail, Suite A   McKinney TX 75070
P: (972)924-5310     W: www.MicroscopyMarket.com

The MIP:  Microscopy & Spectroscopy Business Specialists
Bending the curve toward faster sales and business development
At 08:22 AM 7/13/2015, you wrote:

>*****
>To join, leave or search the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>Post images on http://www.imgur.com and include the link in your posting.
>*****
>
>Hi Steffen,
>Good question! The factor 0.61 comes from the calculation of the intensity
>distribution of light after the diffraction at circular apertures.
>Here is a link in german
>http://www.chemgapedia.de/vsengine/vlu/vsc/de/ph/14/ep/einfuehrung/wellenop
>tik/interferenz2e.vlu.html
>Best
>Jo
>
>ETH Zürich
>Joachim Hehl
>Staff Scientist
>ScopeM - Scientific Center for Optical and Electron Microscopy
>HPM E 14
>Otto-Stern-Weg 3
>CH-8093, Zurich
>Switzerland
>
>
>Web: www.lmc.ethz.ch
>Phone:     +41 44 633 6202
>Natel:     +41 44 658 1679
>Fax:       +41 44 632 1298
>e-mail: Joachim.Hehl <mailto:[hidden email]>@scopem.ethz.ch
>
>
>
>
>
>
>
>
>On 7/13/15 11:26 AM, "Steffen Dietzel" <[hidden email]> wrote:
>
> >*****
> >To join, leave or search the confocal microscopy listserv, go to:
> >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >Post images on http://www.imgur.com and include the link in your posting.
> >*****
> >
> >Dear all,
> >
> >two questions, first about the formula:
> >
> >I thought 0.61*lambda/(NA)  (Rayleigh) would be the appropriate formula
> >for self-luminous objects (dark field, fluorescence) while for bright
> >field the fitting variant of the Abbe formula should be applied:
> >lambda/NA-obj for central illumination, lambda/2NA for NA-obj = NA cond
> >and, in general, lambda/(NA-obj + NA cond), provided that NA-obj is
> >equal or bigger than NA-cond.
> >
> >It seems though, that others are happy with 1.2*lambda/(NA-obj + NA
> >cond). Where does the 1.2 (or 2*0.6)  in a bright field situation come
> >from? Would that be the paper by Hopkins and Barham (1950), that Mike
> >was referring to? Because I have no hope of understanding this paper by
> >reading through during this life (Lots of formulas my biologist's brain
> >is not flexible enough to adjust for).
> >
> >Second:
> >Is there a good source for diatoms that can be stained with fluorescent
> >dyes and self-mounted, as George suggested? I have a few diatom slides
> >for teaching (Pleurosigma), but they came already mounted.
> >
> >Steffen
> >
> >>
> >>
> >> --
> >> ------------------------------------------------------------
> >> Steffen Dietzel, PD Dr. rer. nat
> >> Ludwig-Maximilians-Universität München
> >> Walter-Brendel-Zentrum für experimentelle Medizin (WBex)
> >> Head of light microscopy
> >>
> >> Marchioninistr. 27
> >> D-81377 München
> >> Germany