Hi Holly -

There should be no problem with running Metamorph and Micromanager
simultaneously. With very rare exceptions, Micro-manager will run with
the standard drivers provided with the manufacturer. Even if it does
require a special driver, this would have to be installed separately,
and so installing Micro-manager won't touch any of the drivers you are
currently using.

We have a Nikon spinning disk system running both NIS-Elements and
Micro-manager, and we've run several Andor systems under Micro-manager
as well, so I don't think this should be a problem.

Feel free to contact me directly if you want more advice or help with
getting this set up.

Best,
Kurt

On 9/30/2015 10:35 AM, Holly L. AARON wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear All -
>
> We have been struggling (for some time) with a Nikon/Andor spinning disk
> system running Metamorph and I am now wondering about trying MicroManager
> on this system. It is in regular use, so we cannot afford to disrupt the
> users who are not having issues. One major problem right now is with the
> PFS and a difficult sample with a matrix gel and multiple well-plates.
>
> Anyway, my question to the experts out there is whether or not we can run
> uManager simultaneously on a system running Metamorph or if they will
> interfere with each other and we should not bother?
>
> If they won't play well together, I would be interested to know if there is
> a Metamorph User Community out there? Please point me the way! We are not
> able to get any support from any of our vendors on this software and feel a
> bit stuck with it now.
>
> Thanks!
> -Holly
> __________________________________________________________________
> Holly L. Aaron
> Molecular Imaging Center, Cancer Research Laboratory
> University of California, Berkeley
> 251 Life Sciences Addition, MC #2751
> Berkeley, CA  94720-2751
> 510.642.2901 office      510.426.5005 GoogleVoice
> [hidden email]      http://imaging.berkeley.edu
> Also:
> 444-B Li Ka Shing Center for Biomedical and Health Sciences
>


--
Kurt Thorn
Associate Professor
Director, Nikon Imaging Center
http://thornlab.ucsf.edu/
http://nic.ucsf.edu/blog/