Re: reflected light profilometry trrough plastic - mystery bands
Posted by
Steffen Dietzel on
URL: http://confocal-microscopy-list.275.s1.nabble.com/reflected-light-profilometry-trrough-plastic-mystery-bands-tp7584686p7584694.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
From your image it seems much easier to estimate the thickness of the
spacer rather than the one of the air filled chamber. Not surprising
considering the mess in optical properties you have in the air chamber
light path in combination with the high NA air objective (see Guy's
post). Is spacer thickness not good enough?
Also, if the low NA objective approach Guy suggested should not work
out, maybe you can align the mylar with the polarization of the laser
beam - what happens if you turn the sample?
Steffen
--
------------------------------------------------------------