Hi Zdenek,
the start and end of the z-scan was within the top and biottom mylar
sheets, so you do not see the reflections from the outside surface of the
sandwich.
I was only interested in the channel itself, not the exact thickness if the
Mylar sheets - I checked and they are each about 170 um thick.
The reflections off the spacer show about 40 um apparent thickness, so
with correction for RI it is abour 60 um.
The last reflection off the air channel (4th peak in the profile) corresponds
to 60 um depth from the first reflection, so that seems to make sense and
agree with the spacer thickness, but the two reflections in between are the
puzzle. The image shows the third reflection in the air channel (at 40 um)
that happens to be ain the same position as the bottom reflection of
spacer.
I will try few more tricks, probably wwill end up filling the channel with
fluorescent dye solution. having a less confusing images for analysis will
be a plus.
Stan
.
On Sun, 31 Jan 2016 04:35:26 +0100, Zdenek Svindrych wrote:
>Hi Stanislav,
I would expect two reflections outside the channel ([air - first mylar] and
[last mylar - air]) and four reflections across the channel (every air -
mylar inerface), unless your whole sandwich is glued to some glass support
or something...
This does not quite fit your measured profile, as the two reflection
(outside the channel) should be 360 um apart (without correction).
Furthermore, the [channel air - mylar] interface further from the objective
should not correspond (in depth) to any reflection outside the channel, the
air channel should appear as having different thickness than the mylar
spacer, due to different depth correction.
I can't see any of those in your images... Is your channel really air-
filled? Can you fill it with fluorescent solution and take a profile of the
fluorescence alongside with the reflection?
Best, zdenek
--
Zdenek Svindrych, Ph.D.
W.M. Keck Center for Cellular Imaging (PLSB 003)
University of Virginia, Charlottesville, VA
http://www.kcci.virginia.edu/tel: 434-982-4869
---------- Původnà zpráva ----------
Od: Stanislav Vitha <
[hidden email]>
Komu:
[hidden email]
Datum: 29. 1. 2016 15:51:19
Předmět: reflected light profilometry trrough plastic - mystery bands
"*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Hallo,
I have a little mystery.
I am trying to measure depth of a microchannel (empty, air filled) using
confocal microscopy in reflected mode.
The microchannel is created by sandwiching a ~20 um PET (Mylar) spacer
between two ~170 um thick Mylar sheets. The spacer has a thin layer of
cyanoacrylate glue on both the top and bottom side.
I am performing line-Z scan, with 0.25 um z-step, using a 20x/0.75 dry
objective.
I expected to get two reflections, one from the top of the channel, one
from the bottom of the channel, just like it worked on microchannels made
out of glass that I imaged previously.
However, I am getting additional reflections and am not sure why and also
do not know which one I should take as representing the bottom of the
channel.
See the image here:
http://i.imgur.com/Scj4i6y.jpgReractive index of mylar is about 1.5, but I do not think I need to correct
for z-axis compression since the channel is fille with air.
The reflections are at 20, 40, and 60 um from the first reflection (top of
the
channel)
The lineZ-scan through the spacer shows two reflection peaks, nominally
~40um apart, after z-axis correction it is ~60 um. Even if this is true and
the spacer is really 60 um thick, I am still puzzled by the extra
reflections
in the air channel.
Any hints/input is appreciated.
With regards,
Stan Vitha
Microcopy and Imaging Cneter
Texas A&M University"